Fault:Have a Keithley 480 Picoammeter I pulled off the pile for repair, the fault is when powered on, no display.
Troubleshooting:Opened it up and checked the
instruction/service manual and performed the power supply checks on Table 6-2. All voltages present except step 6, the +5V supply.
On this version PCB (will get the board version later) it has tantalum caps namely C115 on the input of the +5V regulator VR103 and C116 on the output of that regulator, so was immediately suspicious (due to the devices age).
I confirmed (as it came with no battery option) that the internal switch is on the LINE position. Saw
Eevblog #406 which works when set to line voltage as well, also noticed the +5V supply is using electrolytics on that revision, Daves is a black case, this one is a beige case.
I confirmed voltage measured approx 10-11V at TP2 and was present at the leg of voltage regulator input, it needs at least 7V to account for the drop.
I disconnected everything, discharged the axial caps then tested the resistance between ground and the +5V supply (output of VR103). Easy to identify as on the underside of the PCB there is a noticeable star ground. Approx 500ohms which I thought could be low, I didn't notice any capacitor charging either.
Isolating the fault:I removed the +5V supplies output tantalum C116, measured again between ground and the +5V and got 1.4k ohms (getting better?). The tantalum measured about 20% low on capacitance out of circuit, predictable but I didn't get around to testing the ESR.
Checked again and still no +5V present on the output of the VR103 regulator so I removed that as well. Again 1.4k ohms present on the output side to ground (both ways) with it removed. Then I tested VR103 out of circuit at 9V on a current limited supply, it shows good so far and measured +5V on the output. But I did not get a chance to load test it.
I ran a diode test in both directions and discovered 0.5V drop positive to ground and approx 1V drop the other way. So this along with the resistance has got me curious.
Since no battery is present I think I can ignore U301 as that is on pin 1 of the battery connector and I confirmed this with continuity test.
Looking around on the +5 supply rails...
+5V to Q102 Q104 driving the U104 opamp for the front end zero adjustment
+5V to pin 13? and 14 on NAND IC of the clock circuit?
+5V to pin 1 and 4 of U101 the processor
+5V to Q101 on the D5 line to the display module
+5V to pin 3 and 16 of U202 on the display module
I have removed the display module when testing but I didn't confirm what exactly was isolated.
What should I do next:A. There is a low ohms short still present. Remove the ....... next.
B. Apply a 5V current limited to ..... mA onto the PCB where the output of VR103 was.
C. Test the C116 tantalum caps ESR and try to reform it if possible.
D. I believe the C116 tantalum cap was the problem just replace it.
E. Test the +5V regulator VR103 with a load.
F. Just replace that regulator it's dead.
G. Do this....
More info needed:If you have a Keithley 480 can you please take a peak and tell me what voltage and capacitance electrolytics you have installed for C115, C116 and what VR103 regulator you have installed? Let me know if it's been repaired.