Electronics > Repair

INSTEK spectrum analyzer GSP-827 -6 dBm down than the real value

<< < (13/13)

SM5DIV:
Thank you for submitting the manual.
Will now do some in the various registers in the GSP-827 that have had to wait since the device drifted away, over time, concerning the accuracy of various readings.

Addition:
Tried LDCode.exe which won't connect to DUT. All RS232 pins and comport are completely ok. In DUT only 57600 bps is available so I assume the program is preset for this. In an RS232 breaker I see that the program asks for a few ms but the DUT is completely silent. Can anyone give more tips?

Late December:
Got all RS232 communication up and running. Works perfectly.
After calibrating all 6 positions in the Damper together with the linearization module, all input values ​​are now correct.