Thanks, guys!
Please see my response below.
Have you tried throwing an Ammeter in the loop to double check everything makes sense?
No, but the same insulation testers show 10/11 MOhms input impedance for other meters I have, including a "blind" (i.e. non-working OLED display) Agilent U1253A.
So, I doubt it is the insulation testers at fault here.
See note (1) on page 153 of the manual.
https://docs.rs-online.com/071f/0900766b80dc904a.pdf
I am not sure what to make of that note.
What exactly do they mean?
In any case, testing with a very similat DMM, Agilent U1253A, the input impedance measured by the insulation testers is 10/11 MOhms.
Not excluded, but I think it is unlikely that Agilent/Keysight have changed much on the input between the U125xA and the U125xB models.
If you manually select the 1000V range and then test with another DMM, what do you get?
I get 10 MOhms impedance.
What do the insulation testers indicate in the AC and AC+DC ranges?
In AC mode, the measured input impedance is OL for all ranges, which makes sense given the capacitor in the circuit for that measurement mode.
In AC+DC mode, the measured input impedance is 5 MOhms for all ranges.
So, it must be something in the DC mode path.
What is resistance accuracy of used insulation testers? 5-10MOhm is considered low resistance for such meters.
I don't have the specs at hand, but it's not the resolution or accuracy of the insulation testers.
What is the input resistance of other DMM tested with insulation testers?
With other DMMs but same insulation testers, the measured input impedance is the typical 10/11 MOhms.