I suspect the primary winding of the Inverter Transformer in my Tek 2230 is shorted. When measuring its resistance between terminals 1 & 2 with my DVM, I find 0,7R. What do you think ?

Part number is 120-1439-01 (photo attached). In the worst situation, is this transformer on the unobtainable parts list ?

Thanks !
 

Offline TheMG

  • Frequent Contributor
  • **
  • Posts: 866
  • Country: ca
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #1 on: December 08, 2022, 02:16:31 pm »
Not impossible but rather unlikely.

A low DC resistance is normal for such transformers. Since they operate at high frequencies the number of turns is small when compared to a 50/60Hz transformer, so the DC resistance being less than 1 ohm is actually not unexpected.
 
The following users thanked this post: wn1fju, THM60

Offline WattsThat

  • Frequent Contributor
  • **
  • Posts: 766
  • Country: us
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #2 on: December 08, 2022, 02:36:49 pm »
<1 ohm is expected, not shorted. Do waveforms 42/43/44/45 match what is shown in the service manual?
 
The following users thanked this post: THM60

Offline THM60Topic starter

  • Contributor
  • Posts: 17
  • Country: fr
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #3 on: December 08, 2022, 03:38:49 pm »
Thanks for responses.

<1 ohm is expected, not shorted. Do waveforms 42/43/44/45 match what is shown in the service manual?

I can't check that for the time being as the fuse keeps blowing. Two of the three neon bulbs flash (DS858,DS870) and fan shudders one or two times but scope doesn't turn on. 
« Last Edit: December 08, 2022, 03:44:17 pm by THM60 »
 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 3366
  • Country: fr
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #4 on: December 08, 2022, 07:58:56 pm »
get service manual follow PSU tests and troubleshooting flow chart

j
Jean-Paul  the Internet Dinosaur
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 6995
  • Country: ca
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #5 on: December 08, 2022, 09:51:45 pm »
From my notes, Pre-regulator transformer T906 (120-1439-01) readings:
T906 1,2  3.50mH 2.05 ohms Primary Winding
T906 4,5  3.75mH 1.57 ohms Reset Winding
T906 9,10 4.50uH 0.22 ohms Tank Winding
T906 6,7  452uH  2.00 ohms Aux Winding
another person's measurements: https://www.eevblog.com/forum/repair/tektronix-2213a-scope-exploding-power-supplies!/msg864495/#msg864495

I say don't run the scope if DS870 (CRT cathode-grid protection neon) is lit steady. You can see polarity by which pole (-) is lit.
I'd suspect the -2kV rail, it might have a heavy load (cooked focus divider resistors) or that output on the HV multiplier (rectifier) has failed where it outputs AC.
 
The following users thanked this post: THM60

Offline THM60Topic starter

  • Contributor
  • Posts: 17
  • Country: fr
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #6 on: December 09, 2022, 01:00:17 pm »
Thank you so much, I'm going to investigate further with the figures you indicated in hand :-+
 

Offline marcumr

  • Newbie
  • Posts: 7
  • Country: it
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #7 on: December 09, 2022, 01:53:25 pm »
Hello.

Check out that yellow tantalum, mine had a hole in the bottom and shorted out
 
The following users thanked this post: THM60

Offline THM60Topic starter

  • Contributor
  • Posts: 17
  • Country: fr
Re: Inverter Transformer 2200 Series Oscilloscopes, primary resistance
« Reply #8 on: December 09, 2022, 02:29:59 pm »
Hi again Floodybust,

I've removed T906 from the main board (pain in the back) for measurements. Inductance and resistance as follow.
 
« Last Edit: December 09, 2022, 04:11:22 pm by THM60 »
 