I just got a Keithley 199 for 40$ and although the body is a bit roughed up, the face which was covered in sticker residue (including the actual display area...

?) cleaned up nicely with 99% alcohol and now looks new without even a scratch.

I have started to compare it to my other meters to get a feel for it's calibration state and so far DCV and DCA are very good for the ranges I can test. (For example, after appropriate warm-up time etc. the 2.5V output of an AD584 reference reads 2.49945 vs my calibrated Keithley 196 which reads 2.499484, 10V out reads 10.0001 vs 10.00036 on the 196). Those are the ranges I use the most by far so I am quite satisfied.
DCA on the 3A scale is also very close to the 196 so no complaints. I got this for the superb display and don't intend to make any critical measurement with it.
But there is a problem with the ohms.
I picked a series of 1% resistors in 18, 180, 1.8K, 18K, 180K and 1.8Mohm values and measured them with several meters.
I compared the 199 to the average of all my other good meters, and to the 196 alone.
The 199 always reads high and is off by 8%, 0.3%, 3%, 4.5%, 5.5% and 6.8% on the different ranges.
That's a lot, but what is most alarming, is that the readings on the 199 never settle. Even on the 196, they will settle after several seconds down to the last digit. On the 199, the last 3 (sometimes 4) digits are never stable and will continue to change forever.
Except when I measure the 1.8M resistor using the 30Mohm range. Then the value is bang on (within 0.012% of the 196) and stable.
What could be the cause of this?
****** EDIT on 2019-03-20 ******
In the end my choice of 18 series resistor for the original test was very bad, since the calibration point for this unit are at 190, 1.9K, 19K etc. Thus being very close the errors didn't look too bad and I was more concerned with the jumpiness. It turned out all ranges were WAY out of wack except 30M and 300M.
In the end I replaced transistors Q16 BJT NPN with a 3904 (it was leaky). This fixed the 300, 3K and 30K ohms ranges.
I also ended up replacing Q11 N-Channel JFET with a pn4901 (failed short, all three pins reading the same 4.5+ Volts when it should have been closed and gate was receiving -14.8 V control signal). This fixed the 300K and 3M ranges.
I also had to recalibrate all ohms ranges except 30M and 300M since they had been calibrated (some by me) with the faults.
*************************************
It was a fun repair and I owe my thanks to Kleinstein.