Author Topic: Keithley 2000  (Read 1918 times)
Offline klaus11Topic starter
Keithley 2000
« on: March 21, 2016, 11:07:43 am »
I'm trying to repair a K2000, fault 600.1, 600.2, 601.1, 601.2. I have replaced all the transistors marked with 6G mmbf4393 but I find with 3 transistors 132 marked:
2 transistors, pins 2 and 3 in short and third transistor Pins 2 and 3 not short.
Like, you can be a double diode NXP.
Someone help me with device 132.
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
Offline RobK_NL
Re: Keithley 2000
« Reply #1 on: March 21, 2016, 01:31:10 pm »
Tell us what problem you want to solve, not what solution you're having problems with
Re: Keithley 2000
« Reply #2 on: March 21, 2016, 01:40:31 pm »
Yes, the scheme marking all snj132199. But all device 6G, exception 3 device marked 132.
I have replaced 3 mmbf4393 device 132, and have general error in DCV
Re: Keithley 2000
« Reply #3 on: March 21, 2016, 04:55:15 pm »
I cleaned device 132, now 3 same measures. Become welded in place.
Surprise all bugs fixed 60X.X. 6G also replaced all transistor.
30X.X. errors appear
  Before they were not. You can mistake buffer masked by Q109, Q106, Q105 ...