Just in case, very quickly measured resistances (mains just via mains, switching selector via all values), transformer has revision B:
Primary
64.70 White Blue
80.45 White Blue Violet
241.85 White Blue Violet Gray
282.97 White Blue Violet Gray Red
White Blue 64.70
Blue Violet 15.75
Violet Gray 161.40
Gray Red 41.12
Secondaries
Brown Brown 0.757
Yellow Yellow 58.870
Green Green 1.335
Gray/red Orange 5.240
Gray/red Orange 5.316
Mains winding consists of 4 sections, all connected in series, White is always connected to mains, selector then selects from the reset - Blue 100V, or Violet (+20 = 120V), or Gray (+20+100 = 220V), or Red - (+20+100+20 = 240V).
Maybe when I measured voltages I had 220V selector and 230+ in the socket, don't remember, but these voltages are always much higher than rails that we have (and in some cases they are used - if you look in reversed schematics you could find +20V tap going before +15V regulator), that's one of reasons these regulators are hot, and then caps leak ...
With this thread my wish to make single detailed post seems to fail ...
Edited: fixed the latest resistance, accidentally put 2+ ohms