Author Topic: Keithley 2000 display repair  (Read 3726 times)
Offline aixTopic starter
« on: December 22, 2018, 02:15:45 pm »
The display on my Keithley 2000 has stopped working (stays completely blank). I would appreciate some advice on how best to troubleshoot this.
Here is what I have learned so far:
  • Except for the display, the meter is functioning just fine: I can take readings over GPIB.
  • The keyboard (which is part of the same assembly and controlled by the same MC68HC705C8A) also seems to work: the unit reacts to key presses as it would normally: relays click etc. I haven't tested the keyboard beyond pressing a couple of buttons and listening to how the unit reacts.
  • The +37V rail that powers the VFD reads close to zero volts. (Pins 2/3/6/7 of U101: LM317M.)
  • When I disconnect the display assembly from the main PCB, the output of U101 goes to +36.5V and can supply at least 200mA (as tested with a DC load).
  • Supplying +37V directly to the display assembly from a bench power supply puts the power supply into current-limiting mode (at 200mA).
This leads me to the high-voltage part of the display assembly. It consists of a pair of HV5812 drivers and a VFD. There's also a pair of diodes on the back of the board (CR401 and CR402) whose purpose I don't know but that test short (in-circuit, so not sure what to make of this).
If you were in my shoes, what would be your next troubleshooting step?
Thanks for any suggestions and happy holidays!
Offline coromonadalix
Re: Keithley 2000 display repair
« Reply #1 on: December 22, 2018, 02:26:34 pm »
Check your tantalum cap (s)  if there is more
I would have desoldered the hv5812 one after the other by hot air to see if the shorts goes away with one or the two removed ??
I dont think the 68hc705 is short since its 5 vdc powered ??
Here in canada they dont cost too much (hv5812),  under 5$ cad
« Last Edit: December 22, 2018, 02:28:06 pm by coromonadalix »
« Reply #2 on: December 22, 2018, 02:35:26 pm »
I too think that the 68HC705 is fine since it can read the keyboard and talk to the main board. I'm pretty sure it doesn't use the +37V rail, although weirdly the +37V supply is routed to that part of the board (for reasons I haven't quite figured out).
My inclination was to remove/replace the HV5812s (I already have spares on hand). I am just a bit unsure whether my hot air rework skills are up for the job, given how close the glass of the VFD is to the two chips. Need to figure out a heat barrier of some sort?
Checking the tantalum cap is definitely a good idea, thanks. Will do that next and report back.
« Reply #3 on: December 22, 2018, 03:02:46 pm »
The tantalum (C413) is fine (measures 21μF, 0.3Ω ESR).
So I'm guessing the next step is to remove the HV5812. Any practical advice on how to protect the screen and the ribbon cable from heat? (Or any suggestions for other things to try first?)
Thanks again.
« Reply #4 on: December 22, 2018, 05:11:12 pm »
Some aluminum foild can deflect the heat ??  2-3 sheet thick would do fine, or Kapton tape
But the best would be to unsloder the vfd and remove it completly  .... my 2 cents   just to be sure you wont affect it, scratch it, and have plenty of room to work
If youre not to experienced, maybe someone here near your location with the right equipment could do it ??
I would have done for free to help you, but where worlds appart  loll
« Last Edit: December 22, 2018, 05:36:48 pm by coromonadalix »
Offline martinr33
« Reply #5 on: December 22, 2018, 09:12:09 pm »
You could remove the VFD first - easier than pulling the chip. It would be good to eliminate some kind of in-glass short.
Once it is off, if the 37V is still bad, you can pull the high side driver chip.
If the 37V goes good, time to suspect the glass.
« Reply #6 on: December 22, 2018, 09:32:21 pm »
Anyone got any tips for how to remove the VFD without destroying it?
As far as I can tell, it's sitting on double-sided tape. Use dental floss to cut through that and then wick off the solder from the leads?
Offline Rasz
    • My random blog.
« Reply #7 on: December 22, 2018, 09:50:12 pm »
are you serious? :o
hotair: desolder ribbon cable, put a barrier (foil, knife blade, piece of metal sheet etc) between chips and vfd and take them off
Offline martinr33
« Reply #8 on: December 22, 2018, 10:13:13 pm »
Use a solder sucker on the pins. Make sure that evey pin is loose in the hole when you are done by making it move with a small screwdriver. If the pin is loose, it will spring back easily. Sometimes, you have to break them free of vestigial solder with a push. Or you may have to refill the hole with solder and try again.
Once that is done, you can carefully remove the glass. I usually lever them off slowly.
Offline Inverted18650
    • Test Gear Addicts
« Reply #9 on: December 23, 2018, 12:22:37 am »
when I pulled the VFD on a couple, I put a dab of fresh solder on each pin first then used a solder sucker as described above.
« Reply #10 on: December 23, 2018, 01:17:41 pm »
Thanks for your suggestions everyone.
I ended up removing the VFD and both driver ICs, only to ultimately track the problem down to a shorted ceramic cap.  :'(
The display is now working again, but I can't help thinking that there must have been an easier way to track down the root cause. Thermal imaging maybe? Any other ideas?
« Reply #11 on: December 23, 2018, 02:06:25 pm »
Glad it work again,
For trouble shooting  when its possible, track down any schematics if available,  do power supply lines checks, checks for shorts in any signals / data lines ...
Get datashets for any ic's / transistors / mosfets  etc ... you see on the pcb,  may partially serve as service manual ...
And sometimes an 0.05$  part can give you lots of troubles  loll
When i troubleshoot,  i check all capacitors types on the pcb, if i get some low or stranges values, i remove them and recheck, and good lcr meter with serial / parallel mode test help a lot.
And sometime you can have luck  loll
Offline Rasz
    • My random blog.
« Reply #12 on: December 23, 2018, 03:16:41 pm »
cold spray, IPA, lips
« Reply #13 on: December 23, 2018, 05:03:41 pm »
Would you mind elaborating on how to use LCR parallel/series mode for this sort of troubleshooting?
Offline martinr33
« Reply #15 on: December 23, 2018, 08:22:05 pm »
It is quite difficult to hunt down a short like that, especially if it only shows up at higher voltages.
Usually, there are too many suspect parts to make it easy. What you can do is measure the voltage across the rails at various points on the board. Where it is lowest, that is where the short is. You can also trace the voltage drop along the circuit - the voltage drop will follow the current path to the fault.
The voltage will be in the millivolt range. You can also use an external supply. The local supply may have enough ripple that you can use a current tracer like the HP 547A, which is a tricky useful piece of gear if you can find one.
https://www.eevblog.com/forum/reviews/hp-547a-current-tracer/
There are also instruments for tracking shorts - usually an audio osillator that responds to the resistance between points. As the frequency increases, you are approaching the short.