From the service manual:
"Test 407.1 – Front end; 2V range
Cannot measure 7V ±0.15V at A/D IN
The signal at BUFF is routed through R233, K503, K500, the input protection circuit and
Q513 to the input buffer (Q512 and U520). The buffered signal is tied to AMP IN.
The signal at AMP IN is then routed to the variable gain amplifier (VGA) and, being
positive, follows the non-inverting x1 path that consists of R530, Q509, Q507, U519,
R531, U509, Q501, and U516 to pin 12 of multiplexer U511.
The output (OUT) of the multiplexer is routed through buffer U342 to ACV/A. The signal at ACV/A is routed through U320 (/AC pulled low) and applied to the A/D buffer
(U322), which is configured for x1 gain."
This makes no sense to me since the meter works perfectly in the 2V range, and also there are other tests that utilize the identical parts, but those tests pass (such as 407.2 and 410.1).