Author Topic: Keithley 2001 multimeter not completing boot up.  (Read 984 times)

0 Members and 1 Guest are viewing this topic.

Offline dominitechTopic starter

  • Contributor
  • Posts: 17
  • Country: us
Keithley 2001 multimeter not completing boot up.
« on: May 11, 2020, 09:25:01 pm »
I have a Keithley 2001 DMM that has been dropped during transit after getting it calibrated and now the unit will not complete the boot up cycle.  There is fan spin on the rear of the unit so the power supply seems to work but nothing on the display.  After a quick peak inside nothing seems out of place or even busted.  Any tips on what components are susceptible to hard drops or where to check first would be appreciated(Already checked and reseated all cable connections).
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5853
  • Country: ca
Re: Keithley 2001 multimeter not completing boot up.
« Reply #1 on: May 12, 2020, 12:52:41 am »
Any component whos physically big or massive could create problems, cracks in pcb ? connectors broken or have loosen up ?
 
The following users thanked this post: dominitech

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Keithley 2001 multimeter not completing boot up.
« Reply #2 on: May 12, 2020, 01:04:54 am »
99.9% sure display glass is broken.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: dominitech

Offline dominitechTopic starter

  • Contributor
  • Posts: 17
  • Country: us
Re: Keithley 2001 multimeter not completing boot up.
« Reply #3 on: May 12, 2020, 03:20:24 am »
Most of the damage appears to be on the rear.  The aluminum frame is very warped on the back but thanks, I will check the display next.
 

Offline dominitechTopic starter

  • Contributor
  • Posts: 17
  • Country: us
Re: Keithley 2001 multimeter not completing boot up.
« Reply #4 on: May 12, 2020, 03:30:28 am »
99.9% sure display glass is broken.

The display was indeed definitely damaged!  Thanks again!
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5853
  • Country: ca
Re: Keithley 2001 multimeter not completing boot up.
« Reply #5 on: May 12, 2020, 10:41:50 am »
Oh  damn
 

Offline dominitechTopic starter

  • Contributor
  • Posts: 17
  • Country: us
Re: Keithley 2001 multimeter not completing boot up.
« Reply #6 on: June 05, 2020, 10:18:35 pm »
Do you suppose I can use a display from a Keithley 2000?  The part numbers slightly differ from each other and Keithley will not sell me a new one.
 

Offline picburner

  • Frequent Contributor
  • **
  • Posts: 497
  • Country: it
Re: Keithley 2001 multimeter not completing boot up.
« Reply #7 on: June 06, 2020, 04:19:43 am »
The Keithley 2000 vfd is completely different (1 row only). The Keithley 7001 vfd is the ideal donor candidate.
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: coromonadalix


Share me