Hello there, I have here a Keithley 2001 witch show powered on just „no comm link“. I replaced all the old capacitors, erased and reflashed the EPROMS but the problem is still alive.
With the help of the service manual form xdevs.com I checked the voltages of the power supply and recognize a strange behaviour at U103. Pin 2 should have 8 volts but I measure about 1.2 volts.
Should I replace U103(LM317T)?
Also, on the measure points the -/+ 8 VF are not there. At the beginning +8VF is about 7.5 v and -8 is about -5.5v. The longer the DMM is powered on, the voltages change. After 10 minutes it is about at -2,4V and +11V. I tried to understand this behaviour with the schematic but I have no idea where to look exactly or what it could be.

I also discovered a spot which looked like an arc to me. R303 is one resistor before the TLPs. If it was an arc could that crack the TLP U311? I measured the diod of the TLP U311 (Pin 1 and 2) which looks normal compared to the other optocouplers.
TiN wrote in this
https://www.eevblog.com/forum/repair/keithley-2001-meter-went-for-cal-working-arrived-dead/ post what potentially could cause that comm error.
Sometimes, esp on old meters the cable between boards get dodgy, resulting in your comms errors.
Other times DC/AC/DC converter (ERG potted brick or little PCBA on digital board) that powers VFD glass crap out, making CPU to go sick and resulting same comm errors, that happen to units I serviced (more than one)
.
Or ADC board went loose, that will result comm error. Or the TLP opto on analog board went poo, also comm error.
Or the analog board caps leaked their goo, and ruined everything. 
I checked the cable between the boards witch seems ok. ADC board is also not lose.
Can someone tell me how I can check the other compontents?