Hi,
I've just got my hands on a Keithley 2001 reported as failing a load of tests and having a burning smell!
The burning smell appears to have been U316 (LM339D) which had literally blown itself apart. A new LM339, some careful carbon scraping and a couple of small bridge wires and that problem seems to be solved.
From a test standpoint it fails 301.2 and then a large number of the follow-on 300 tests -> 301.2, 302.2 303.2, 304.2-7, 305.2, 307.2, 307.3 and 308.1
On DCV it shows about 340V with no input connected ... connecting a voltage source impacts the voltage by what appears to be the correct amount, i.e a 1V input increases the display to 341V and it seems to be pretty accurate.
DCI appears to work perfectly, so I know the ADC is ok.
I have just taken out the Analog board to replace all the caps (which looked and tested perfectly fine, but I've ordered replacements anyway) and I noticed some strange behaviour on VR302 and VR307, both should be 6.4V temperature compensated zeners (1N4577) ... I wasn't seeing a forward voltage drop, so I've taken both out of circuit and when testing them on my DMM7510 (at 1mA) it shows about 6.2V in the reverse direction, but in excess of 10V in the forward direction ... is there any reason why this might be the case? Normally I'd just suspect a faulty zener, but the reverse voltage is correct, and both zeners are showing the same behaviour.
They are both on a related part of the schematic, both seem to be involved in generating +8VF and -8VF ... which probably not coincidentally are also the rails that feed U316.
But, all of the voltages seemed to be normal when tested, but I don't really understand the bootstrap part of it, so I am very likely missing something.
Any help appreciated,
Lee.