The instrument, like many others in this category, on the lower ranges (100mV - 10V) has an input impedance >10Gohms.
If you leave the inputs connected to nothing it is normal that the last digits move continuously, the instrument captures any disturbance.
If you try to connect it to a stable source you will see that the reading will be equally stable.