Hi guys, thanks for your replies.
Here are the updates:
I tried removing everthing back to the 4094s. Probing each output revealed that one of the 4094s was bad. It would reflect pulses off the input stream directly to the output and have a rather high low-level voltage of about 2.5 V. Changed that immediately. Now all logic levels were correct. I then mounted an LM339(at U115), still without JFETs Q113,115,104,108 and probed its outputs. It worked! For a little bit, that is. I then tried lifting its legs, but it was gone. I then removed that one and mounted a whole new LM339, with all output legs lifted. Same thing, it worked but only a little bit. confuses me even more, cause it means that there is something going on with the digital signals or the power, that caused the IC to die. Simply can't explain it. Probing back, it looks like one of the 4094 outputs is dead as well now. Edit: They all appear to work.
>o too much current in output -- not likely as the outputs seem to drive only FETs.
The LM339s are able to withstand a short to ground(in this case -15)V for an unlimited amount of time. So, in theory at least, that shouldn't be the problem.
> o too much current in input if brought below Vee -- two of the LM339 inputs are driven by signal 'OHMA' but I can't see where it comes from. Vee is apparently tied to AGND so if 'OHMA' is brought below -0.3 V relative to AGND and enough current flows, the LM339 could be damaged.
I haven't even gotten to the OHMA opamp yet. Im basically only focusing on U115, since its outputs are the only ones that do not working according to the switching tables in the service manual.
> Supposing the digital and analog grounds are supposed to be isolated, maybe there's a shorted cap or other component applying a potential difference between the two grounds. Look in the power supply section for caps to ground on the transformer secondary and regulator chip insulators. I recall one instrument where an SMD part fell off the PCB and became a jim-dandy short in another part of the circuit.
There are no power issues that I can see. Everything is fine and stable, no overshoot at startup, no shorts or anything.
> I would hold off changing IC's, there's something else going on. Clean the PCB with IPA and q-tips, look for solder troubles first with a magnifier.
Yes, there is definitely something else going on. Its very weird.
I thought it might be latch-up as there are three rails (+5V, +15V, -15V) involved and if the +5V is late coming up, you might get the LM339 latching up through the CD4094?
Could be, but there must be some coupling effect here. The CD4094 works just fine when U115/LM399 isn't mounted, but as soon as it gets mounted, it starts acting up, possibly taking CD4094's with it.
> Is U120, the second LM339 working OK? No IC's running hot?
Nope. I don't have a thermal camera and I ain't too happy about poking around in there(due to contaminating the sensitive areas), but the digital stuff and the control stuff for switching the JFETs doesn't seem to get hot.
> I can't see a JFET causing any of this, most have gate resistors. It's good to check V across for a leaky/shorted gate.
I completely agree. I have not seen any shorted FETs or similar. At the moment they are all removed and there is little continuity between their pins.
Edit: I dont know what happened, but it seems that its going the right way now. I got Q104 and Q108 working now. I tried swapping out the LM339 *agan* and was probing around, and suddenly it was working. Now, Q105 and Q113 still aren't working, but I discovered that I can get the output that controls Q105 into a pretty severe osillation by changing around the ranges on the multimeter. But I can't tell if the 4094 is dead again, now..
I am starting to suspect that having those JFETs in actually helps the circuit stabilize or something(I know its really floating without, but that shouldn't affect the comparator..). Anyway, I'll do some more swapping tomorrow and see where it takes me.