Author Topic: Keithley 614 Electrometer - Repair [Success]  (Read 6872 times)
Offline HighVoltageTopic starter
« on: June 04, 2016, 10:45:20 am »
This was a great deal, as long as I get it repaired and restored, my first Keithley Electrometer.
Made in the USA
Rev. F
Problem:
- Does only turn on intermittently
- Shows random numbers on the display
- Has a part of the front bezel broken out
This meter is also battery powered, but the batteries are dead
I will first have a look at all the switches, which feel a little sticky.
The front panel has a old style Triax Input connector with two pins like a typical BNC connector and not the new 3 pic triax connector.
Hmm, I wonder if I should exchange the input connector to a more modern type.
But first, lets get it back to work.
Anyone here have experience with this 614 Electrometer?
« Last Edit: June 11, 2016, 01:46:24 pm by HighVoltage »
Re: Keithley 614 Electrometer - Repair and Restauration
« Reply #1 on: June 04, 2016, 12:16:25 pm »
OK, I was so much bothered by the damage on the front bezel plastic part, I fixed it first
Step 1: Clamp two aluminum pieces with double sided tape against the bezel.
Step 2: Fill the slot very slowly with some epoxy resin.
Step 3: Use fine sandpaper to grind it to perfectly matching shape
The color is not completely matching but the shape is perfect.
May be I will have a calibration sticker right there and then nobody will notice.
I am happy with the plastic repair but just not with the color.
« Last Edit: June 04, 2016, 12:24:43 pm by HighVoltage »
The following users thanked this post: yus
« Reply #2 on: June 04, 2016, 12:24:06 pm »
Here are some specifications of this 4.5 digit Keithley 614 Electrometer
- Detecting currents as low as 10^-14 A
- Resistances up to 2 x 10^11 ohms.
- It provides more than 5x10^13 ohms input impedance on the voltage ranges
- Can detect charge over a range of 10^-14 to 2 x 10^-08C.
- Voltage ranges are 0.2 V to 20 V in 3 ranges with accuracies to 0.08%.
- Current ranges from 20 pA to 2000 µA in 9 ranges with accuracies from 1.5 to 0.3%.
- Resistance ranges are from 2 kohms to 200 G-ohms in 9 ranges with accuracies from 0.5 to 2%.
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #3 on: June 04, 2016, 04:10:45 pm »
Beauty. HighVoltage is running ahead, getting first electrometer  :scared:. I should do something about that asap.  :-DMM
Interesting trick with panel repair. Perhaps paint it with matching color later to have it look genuine? :)
For example, repair 41422A SMU module...
Offline vindoline
« Reply #4 on: June 04, 2016, 06:44:25 pm »
HighVoltage, nice job on the plastic repair! What kind of epoxy did you end up using??
Also, you look lucky with the triaxial Jack. It looks easy to replace if you want. I have a Keithly 616 and it looks to be a bear to get to the Jack to change it. The mod I did (and it's a bit of a hack) is to file off one of the lugs on my input jack. I can now use either 2 or 3 lug cables.
Offline Kleinstein
« Reply #5 on: June 04, 2016, 07:45:11 pm »
The 614 manual include schematics, which should make a repair relatively easy. The ADC is a standard ICL7135 - so easy to change if needed. IF the input FETs are broken it might get difficult to find some as good as the original, but at least reasonable performance should be easy to get.
Intermittent turn on points toward the power supply part. Could be trouble with caps or the oscillator. This would be the first part to check anyway.
Offline Smith
« Reply #6 on: June 05, 2016, 07:46:11 am »
I know there is a trick for the 617 input. There are some good and cheap opamps (LMC662) wich offer a beter and more (temp) stable input. Just take a look at https://www.eevblog.com/forum/testgear/fun-with-low-leakagebias-current-femtompere-electrometer-keithley-617/
Btw before replacing the triax with a three lug triax you might want to check for some cables with two lug triax. I know the 3 lug are very expensive, the 2 lug might be a bit less expensive.
« Last Edit: June 05, 2016, 07:48:11 am by Smith »
Trying is the first step towards failure
« Reply #7 on: June 05, 2016, 11:17:07 am »
Some interesting suggestions, Thanks
I might paint the bezel later.
Here are some more pictures.
Display PCB
And the 50 T ohm input arrangement on Teflon standoffs
« Reply #8 on: June 05, 2016, 11:41:00 am »
This is the resin type epoxy I used.
I added some black and little yellow color, to get a matching brown, but that failed.
It is pretty strong attached to the original plastic.
The following users thanked this post: vindoline, yus
Offline Johnny10
« Reply #9 on: June 05, 2016, 03:24:08 pm »
Didn't FeedbackLoop just go over one of these on You tube?
I think problems with switches.
Will have to look at that again.
« Last Edit: June 05, 2016, 03:34:22 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
« Reply #10 on: June 05, 2016, 04:43:14 pm »
Today I cleaned all switches but it did not help
Next is the power supply testing
+/- 5V is ok
But it seems the 32V rail is sitting at 25 Volts hmm
Offline feedback.loop
« Reply #11 on: June 05, 2016, 04:52:29 pm »
That is right. The switches were the culprit in my case. I also fitted a 3-lug triax connector.
Here is the video:
https://youtu.be/txH6EpqN5V0
« Reply #12 on: June 06, 2016, 07:19:57 am »
Thanks Feedback.Loop for the link to your great video.
After cleaning all the switches on my 614 it turns on perfectly but still shows strange and random numbers
Time to get out the schematics and start at the power supply
« Reply #13 on: June 06, 2016, 08:16:45 am »
By total surprise, I found an old original Keithley 2-lug connector and cable in a local store here in town.
I will just install some banana plugs on to the cable and see how good it performs
It will be a good start and I don't have to exchange the connector in the instrument.
« Reply #14 on: June 11, 2016, 01:46:00 pm »
The repair was a success !
The power supply had a bad regulator, not sure why, no obvious reasons are visible and it works perfectly now.
Calibration was easy as well.
But my lowest current source is the Keithley 2450 SMU and it does "only" do 1nA
And the Keithley 614 has a pA range
Hmm, no way to test it.
I think I will hold off on the painting. The repair of the bezel does look ok as it is.
Now I need to build a metal enclosure for high resistance measurements.
I am amazed by this little and old Keithley 614 and its capabilities with 50 Tera Ohm input resistance.
Verification:
1 nA
1 uA
19.0 V
Offline Smith
Re: Keithley 614 Electrometer - Repair [Success]
« Reply #15 on: June 11, 2016, 10:51:29 pm »
One method i found out to be helpfull, in need of some Tohm resistors: use a standard LED as a current source, and take down the light until your range reaches about 10% full scale, then range down one scale. Not absolutely accurate, but good enough for testing. BTW, it does differ a lot depending on your light and the type and color LED. You could  even use a second LED to shine on the first one to make it a bit more stable and controlable, but rember the setup is quite temperature dependant (due to the leds themself).