Hello,
This is my first topic in this forum, I will apreciate help with that, thanks in advance.
I bought a broken 20Mhz Kenwood CS-4025 Oscilloscope and I have set myself the challenge of fixing it!!
The guy who had it broked it plugin a bare wire to the wall 220Vac plug, i suppose he connected the ground to the hot wire.
When I powered it for first time, the green led is on but nothing in the screen. Searching the failure with the service manual i found that the high voltage for the CRT is dead.
For accessing the power supply pcb I needed to take out the CRT, the test points that i write are without the CRT connected.
TP1: ok
TP2: ok
TP3: ok
TP4: ok
TP5: ok
TP6: KO
TP7: KO
TP8: KO (only noise, no 1800V)TP9: ok
TP10: seems ok, but the waveform seem to bo slighly differentTP11: seems KO (having a square waveform with minimum voltage of -50V)TP12: KO (0V)TP13: ok
TP14: ok (but is 15-20V less in voltage)
I attach the service manual:
https://www.vintageshifi.com/repertoire-pdf/pdf/telecharge.php?pdf=Kenwood-CS-4025-HU-Service-Manual.pdfpage 17: High voltage checking procedure
page 28: Power supply schematics and TPs (test points).
Any suggestions to continue?
Many many thanks!!!!!