Author Topic: Keysight / Agilent MSO-X 7000A Repair  (Read 841 times)
Offline smaultreTopic starter
« on: December 26, 2023, 10:20:11 am »
Maybe somebody have any good service manual for it?
It's boot and halt with message PLL unlock, uncalibrated..
Do not response on any buttons.
Start a new life here!!!
Re: Keysight / Agilent MSO-X 7000A Repair
« Reply #1 on: December 27, 2023, 02:10:08 am »
I make some investigations.
There is some measurements, 2 chips marked with RED getting really hot :-\.
Scopes boot to measurements screen with error message:
"software failure,PLL unlock, uncalibrated, defaults loaded"
 Also it do not feed 5V to USB HOST connectors (can't read usb), but
 when i plug it to PC its find a new device..
 U1809 - oscillator powered, but tuned off state (worked, tested)
 U2901 74act04 Hex Inverter getting very hot
 U2520 act11374 Octal D-Type Edge-Triggered Flip-Flop getting very hot too
 500Mhz VCO generator module itself seems to start
but ACQ chips (5 pcs) do not start (no oscillations on pins some have log"1", some log"0")
« Last Edit: December 27, 2023, 08:52:28 am by smaultre »
Offline HighVoltage
« Reply #2 on: December 27, 2023, 11:32:53 am »
This is a MSO 7000A (No X)
If you find a component to be hot and a supply voltage rail low, look for a short.
Either an IC or a capacitor shorted is most likely the fault.
Always start with checking all the voltage supply rails.
Unfortunately there is no good service manual available.
I have repaired a few of these 7000A and 7000B scopes and with enough time and effort, they are repairable,
unless you get a fault in an agilent specific chip.
« Reply #3 on: December 27, 2023, 08:54:57 pm »
Do you have any voltage measurements sheets? Can you test  U1809 - oscillator, does it generating on the default state? Do you know how the VCO works here?
« Last Edit: December 27, 2023, 10:21:08 pm by smaultre »
Offline beyond
« Reply #4 on: December 13, 2024, 01:12:09 pm »
Hello, My DSO7014B cant not start user calibration,  I checked the trigger output buffer (ACT11244) chip input pin and output pin,  no signal, the input pin's routing can't be found after passing through the via, do you konw trigger output signal comes from which chip?
I have described the problem in this thread(latest page), could you please give some advice?
https://www.eevblog.com/forum/testgear/agilent-mso-500060007000-anyone-hacked-these-scopes/700/