Author Topic: Leader Lbo-5825 Oscilloscope  (Read 2460 times)
Offline Alf TorpTopic starter
  • Country: no
« on: May 05, 2019, 12:55:54 pm »
new to the forum, dabbling in electronics and repairing all sorts of stuff. Mechanics, electronics and more.
I've scored a Leader LBO-5825 but I get no life in the crt. No trace or other signs of activity. I've been looking for any documentation on the LBO-5825 that I could find, but so far; zilch. Nada. Not so much as a comma.
Can anyone here help out getting for instance the service manual, owners manual, schematics or anything like that? I'd like to get the bugger working again.
Cheers,
Alf Torp
Norway
« Last Edit: May 05, 2019, 12:57:44 pm by Alf Torp »
Offline Chris56000
Re: Leader Lbo-5825 Oscilloscope
« Reply #1 on: May 05, 2019, 08:08:01 pm »
Hi!
Have you tried (carefully) seeing if the Leader badge is removeable? I'm fairly sure the LBO–5825 is a badged version of something you will find a manual for – it looks very much like something of Iwatsu manufacture to me, and Iwatsu do have schematics and service info in many of their manuals!
Failing that, open up your scope and see if you can get me one or two of the PCB reference numbers!
Chris Williams
It's an enigma that's what it is!! This thing's not fixed because it doesn't want to be fixed!!
Offline pbarton
« Reply #2 on: May 05, 2019, 09:44:48 pm »
Looks like the Leader LBO–5825 has two or three more buttons than a Leader LBO-516.
The Leader LBO-516 Oscilloscope Operating & Service Manuals are available on CD from Artekmedia
https://www.ebay.com/itm/391537865516 and http://artekmanuals.com/
« Last Edit: May 05, 2019, 10:23:49 pm by pbarton »
  • Country: no
« Reply #3 on: May 06, 2019, 06:13:59 pm »
Thanks, I'll check it out!
  • Country: no
« Reply #4 on: May 06, 2019, 06:15:37 pm »
Hi!
Chris Williams
Thanks, Chris!
I've opened it up and there are a lot of Leader branded components in there, so I don't know...
The "main board" is T-33338-P1, then there is a board T-33288-P1. I've found the power board, with the rectifiers and some test points marked with what I assume to be the supposed voltage.
Thus, I've found that the 12V, -12V (TP3 and TP4) is missing. Other test points are also showing erratic readings so I am thinking there is some busted voltage regulator business going on, though I'm having trouble pinpointing them.
  • Country: no
« Reply #5 on: May 07, 2019, 06:10:12 pm »
Anyone know what the D859 is? I've not been able to find anything useful, my google fu is letting me down... This is from the power board.
Offline pbarton
« Reply #6 on: May 07, 2019, 10:20:41 pm »
If you go to https://www.manualslib.com/download/1232936/Leader-Lbo-524.html and click where it says "Download PDF" you will receive a .pdf of the Leader LBO-516 service manual.
There is a lot of commonality between your  Leader LBO–5825 and the Leader LBO-516.
On .pdf page 30 of 34 there is the power supply.
Q1 = 2SD859
The following users thanked this post: Alf Torp
  • Country: no
« Reply #7 on: May 08, 2019, 04:31:27 am »
Q1 = 2SD859
Got it. (your advice)
Found it. (the 2SD859)
Ordered it. (same)
...and a couple of other likely suspects.
I downloaded another service manual for the 516 yesterday, but interestingly, the schematics looked rather different. This one looks very similar indeed, so thanks again  :)