Leader LS1020 Oscilloscope Repair
« on: April 08, 2014, 03:33:48 am »
I picked up a Leader LS1020 Oscilloscope amazingly cheap, $20. It is a cute little scope in pristine excellent condition, except there is no vertical deflection. It sweeps fine, so I guess the inputs are blown--both channels.

Anyone have any advice how to attack it? The only schematics I can find on-line are of the power supplies, so I am going in blind.

Thanks for any advice!
"There are no facts, only interpretations."
--Friederich Nietzsche
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Leader LS1020 Oscilloscope Repair
« Reply #1 on: April 08, 2014, 06:05:09 am »
Just offer a simple but distinct waveform (square-wave, even probe cal output) to one input and follow the signal path with another scope. It is possible one fault has affected both channels.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline TheRuler8510Topic starter

  • Regular Contributor
  • *
  • Posts: 179
  • Country: us
  • RF Engineer
Re: Leader LS1020 Oscilloscope Repair
« Reply #2 on: April 08, 2014, 11:12:43 pm »
Just offer a simple but distinct waveform (square-wave, even probe cal output) to one input and follow the signal path with another scope. It is possible one fault has affected both channels.

Good answer...thanks!