Electronics > Repair
Leader LS1020 Oscilloscope Repair
(1/1)
TheRuler8510:
I picked up a Leader LS1020 Oscilloscope amazingly cheap, $20. It is a cute little scope in pristine excellent condition, except there is no vertical deflection. It sweeps fine, so I guess the inputs are blown--both channels.
Anyone have any advice how to attack it? The only schematics I can find on-line are of the power supplies, so I am going in blind.
Thanks for any advice!
tautech:
Just offer a simple but distinct waveform (square-wave, even probe cal output) to one input and follow the signal path with another scope. It is possible one fault has affected both channels.
TheRuler8510:
--- Quote from: tautech on April 08, 2014, 06:05:09 am ---Just offer a simple but distinct waveform (square-wave, even probe cal output) to one input and follow the signal path with another scope. It is possible one fault has affected both channels.
--- End quote ---
Good answer...thanks!
Navigation
[0] Message Index
Go to full version