Working through an intermittent PS problem exactly like DaJMasta's. W/ an unloaded PS, C93 (ceramic, 390pF, 1kV), gets up to ~140F in about 20 seconds (thermal cam).
Pulled C93 and tested it with the DE-5000 LCR @120Hz = 401pF, D 0.021, Q 59.4, ESR 12K-300K, -89 phase angle. The ESR and Q ar alllll over the place.
I grabbed a new (caveat: new from Aliexpress) 470pF, 1kv ceramic. I tested @120hz = 470pF, D 0.003, Q 135, ESR 20k-50k, -89.5 phase angle. ESR also all over the place. This has to be some kind of measurement error, right?
I've pretty comfortable with using ESR to grade electrolytics, but I've never looked at ceramic caps in the pF range. I'm stumped by this measurement.