Author Topic: [Solved!]Looking for firmware dump for Tek THS3024.(How to repair firmware)  (Read 3891 times)

0 Members and 1 Guest are viewing this topic.

Offline fzabkar

  • Super Contributor
  • ***
  • Posts: 2804
  • Country: au
Re: Looking for firmware dump for Tek THS3024.
« Reply #25 on: June 10, 2024, 03:48:00 pm »
The problem is how to recover the part from $4EC0-$FFFF, which we believe has its serial number and calibration constants written to it.
There must be some whitespace between 0x4EC0 and 0x8000. This is so that a firmware update can "grow". Also, both upper and lower ROMs must be corrupt, but not consistently so, otherwise I would not have been able to reconstruct those 4 x 8KiB blocks.

Code: [Select]
Offset(h) 00   02   04   06   08   0A   0C   0E

00008000  FFFF 0200 FFFF 0000 FFFF 0000 FFFF DC0D
00008010  FFFF 0044 FFFF 0000 FFFF 0000 FFFF 0000
00008020  FFFF 1348 FFFF 7011 FFFF 0100 FFFF 6900
........
0000A000  0004 FFFF C419 FFFF 0100 FFFF AF6C FFFF
0000A010  7600 FFFF 0000 FFFF 0000 FFFF 0000 FFFF
0000A020  0100 FFFF CE00 FFFF 2000 FFFF 5500 FFFF
« Last Edit: June 10, 2024, 05:05:28 pm by fzabkar »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 7001
  • Country: ca
Re: Looking for firmware dump for Tek THS3024.
« Reply #26 on: June 10, 2024, 04:30:16 pm »
can this 2014 FW and flash tool help ?   attached


Firmware v01.02 corrects the problem happened during probe calibration. Before the change, Waveform is not visible during probe calibration when certain languages are selected. Less than 2 divisions are visible due to pop-up window size. Detail …

Firmware | P/N 066136603 | Date: Monday, March 17 2014
 

Offline fzabkar

  • Super Contributor
  • ***
  • Posts: 2804
  • Country: au
Re: Looking for firmware dump for Tek THS3024.
« Reply #27 on: June 10, 2024, 08:34:22 pm »
I examined the firmware payload and extracted the metadata for the DATA BLOCKs. I then looked for "holes", ie areas that are not touched by the update.

For example, bytes 0x20dbc2 - 0x20dbc3 are untouched.

Code: [Select]
,#H4020c000,#H1000
,#H4020d000,#Hbc2
,#H4020dbc4,#H1000

This is the corresponding area in dead.bin:

Code: [Select]
Offset(h) 00   02   04   06   08   0A   0C   0E

0020DBB0  0000 0000 0000 0000 0000 0000 0000 0000
0020DBC0  0000 FFFF 0000 4902 0000 0944 0000 0000
               ^^^^
0020DBD0  0000 0000 0000 0000 0000 0000 0000 0000

Is this the original data, or is it now corrupt? Was this area ever programmed, in which case 0xFFFF would be the normal erased state of this word? Have I introduced a red herring?
« Last Edit: June 10, 2024, 08:39:33 pm by fzabkar »
 

Offline fzabkar

  • Super Contributor
  • ***
  • Posts: 2804
  • Country: au
Re: Looking for firmware dump for Tek THS3024.
« Reply #28 on: June 11, 2024, 04:34:14 pm »
Ruby.ldf

Code: [Select]
,#H40000000,#H28
,#H40000028,#H4
,#H40000030,#H8
,#H40000040,#H30
,#H40000080,#H1000

dead.bin

Code: [Select]
Offset(h) 00   02   04   06   08   0A   0C   0E

00000000  1122 3344 5566 7788 5555 5555 AAAA AAAA
00000010  3333 3333 CCCC CCCC 0F0F 0F0F F0F0 F0F0
00000020  FF00 FF00 00FF 00FF 3000 0040 FFFF FFFF
00000030  4000 0040 5800 0040 FFFF FFFF FFFF FFFF
00000040  0080 0040 0040 0000 00C0 0040 0040 0000
00000050  0000 0000 0000 0000 0000 3F40 0040 0000
00000060  0040 3F40 0040 0000 0000 0000 0000 0000
00000070  0000 0000 0200 0000 0020 0000 0200 0000

These words have not been reprogrammed by the update. Are they corrupt?

Code: [Select]
Offset(h) 00   02   04   06   08   0A   0C   0E

00000020                                FFFF FFFF
00000030                      FFFF FFFF FFFF FFFF
........
00000070  0000 0000 0200 0000 0020 0000 0200 0000
 

Online squadchannelTopic starter

  • Frequent Contributor
  • **
  • Posts: 423
  • Country: jp
  • deepl translate user
Re: Looking for firmware dump for Tek THS3024.
« Reply #29 on: June 13, 2024, 07:03:22 am »
omgggggggggggggggggggggggggggg :) :) :) :) :) :) :) :) :) :) :) :)
I will describe the details later, but it is now recognized when I rewrote it appropriately. I am extremely happy!
We will summarize later, as we do not know how the changes we made in the binary affect it.
We will leave the binary that we were able to launch. It is not the final binary, so please do not try it with your own scope.

I haven't had lunch, so I'm going to go eat.

 :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+ :-+


« Last Edit: June 13, 2024, 07:14:55 am by squadchannel »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 7001
  • Country: ca
Re: Looking for firmware dump for Tek THS3024.
« Reply #30 on: June 13, 2024, 10:01:24 am »
congratz   

and yes  if you can  tell us  your latest recovery story ... to help others in the future    :-+
 

Offline fzabkar

  • Super Contributor
  • ***
  • Posts: 2804
  • Country: au
Re: Looking for firmware dump for Tek THS3024.
« Reply #31 on: June 13, 2024, 10:39:07 pm »
Congratulations from me, also. For a while it was looking like you were at a dead end.

That looks like a very nice unit (I want one, too). Let's hope there are no other problems. :-)

https://assets.rs-online.com/v1698844007/Datasheets/4f5c6ab212d4cf09e8376a595b77b528.pdf
« Last Edit: June 13, 2024, 10:41:18 pm by fzabkar »
 

Online squadchannelTopic starter

  • Frequent Contributor
  • **
  • Posts: 423
  • Country: jp
  • deepl translate user
Re: Looking for firmware dump for Tek THS3024.
« Reply #32 on: June 14, 2024, 04:43:38 pm »
I have used it several times and it seems to work fine.
For now, the firmware build is complete. Thanks to all those who helped.

Here is a summary of how to create the firmware.
Although unconfirmed, I believe the same method can be used to restore the Fluke 190 Series II (190-xxx).

Please make sure you have created a directory to do the work. Alphanumeric, of course. Do not use spaces.
  • Dump the firmware from the main unit. There are two FlashROMs that store the firmware, and the firmware is divided and written to them.
    We call them Upper/Lower with the BNC input on top. Both can be dumped using the TL866 and an adapter.

  • Combine the binaries dumped from ROM into one. We have created a "2to1.exe" to make this easy. The source file is also attached.

    2to1.exe upper.bin lower.bin

    Drag and drop Upper/Lower into 2to1.exe to generate an integrated rom.bin.
    Be sure to check the contents. When you open the binary, from the beginning it should be 11 22 33 44 55 .....
    If 33 44 11 22 .... then the Upper/Lower paths are reversed.


  • Download the file for firmware update from the official website (Fluke/Tek).
    In case of Tek THS3000, you can download from https://www.tek.com/ja/support/software/firmware/ths3014-software-1(THS3000_FW_v0102_Installer.zip).
  • Extract only "Ruby.ldf" in the downloaded updater. In the case of Fluke, it is "Tetra.ldf".

    5. Download "FlukeFW2Bin.exe" which creates bootable firmware from the update file. You can download it from the following topic
    https://www.eevblog.com/forum/microcontrollers/fluke-19xbcii-firmware-to-binary-converter/msg1280562
  • Drag and drop "Ruby.ldf" or "Tetra.ldf" into "FlukeFW2Bin.exe" and run it.

    FlukeFW2Bin updateFile.ldf

    After execution, "Ruby.ldf.DataBlock.bin" and "Ruby.ldf.ExtensionData.bin" will be created directly under the directory.
    ExtensionData.bin" is unnecessary. You may delete it.
    Only "Ruby.ldf.DataBlock.bin" is used.
  • Now comes the important part. Integrate the corrupt firmware "rom.bin" and the generated firmware "Ruby.ldf.DataBlock.bin".
    Calibration, serial number, and model name, all three of which are stored in $8000-$FFFF. This range can be further divided into four parts
    $8000-$9FFF
    $A000-$BFFF
    $C000-$DFFF
    $E000-$FFFF
    The blocks are divided into If $8000-$9FFF contains the "correct binary", the calibration, serial, and model name are recognized correctly.
    If not correct, in my case (THS3024), the model name is "THS3014", the serial number is "SERIAL NUMBER", and the calibration is "Invalid".

    For correct recognition, there is a checksum at the end of each block. The checksum values must be matched correctly.
    There are two types of checksums,
    The result of adding the bytes from $0000 to $1FF7 (in the binary range of the block) goes into $1FF8,
    The result of XORing the bytes from $0000 to $1FFB (in the binary range of the block) (including the result of the addition) is placed in $1FFC.

    checkSUM.exe" and "checkXOR.exe" have been created to perform the checksum calculation. The source is also available.
    Extract the four blocks, create a binary, and drag and drop it. The result will be output.

    Check the checksums and if they match, congratulations. You can proceed to the next step.

    If this calculation does not match, it may be damaged somewhere. In my case it was corrupted.
    If it is corrupt, it can be restored based on information from other blocks, but there is a condition: it can only be restored between blocks of $8000 and $A000, and between blocks of $C000 and $E000. (Probably)
    Also, the XOR checksums in the $C000 and $E000 blocks do not match. We believe the calculation method is probably different.
    We have also checked the case where there are only two blocks ($8000-$BFFF and $C000-$FFFF). In this case also, the XOR checksum results do not match.
  • Change one more location. This is the $0070-$007F part.
    This part can be filled with 00h or copied from a corrupted ROM. Either way, it worked fine.
    In the generated firmware, there are "FE" and "F4" at the beginning of the area, but this did not work.
  • Ruby.ldf.DataBlock.bin", which was integrated in steps 7 and 8, needs to be split into two parts for writing to the flash ROM.
    We created "1to2.exe". Source available.

    1to2.exe rom.bin

    Drag and drop the merged "Ruby.ldf.DataBlock.bin" and it will be split into upper.bin and lower.bin.
    There is no confirmation that the files will be overwritten, so it is better to create a new folder and use it.
  • Write the upper/lower created by the split, solder it, and check if it works.

The above is how I have done it with success.

There is a lot of mystery about how the checksums are calculated. In my case, the XOR checksum did not match for the $C000 and $E000 blocks.
After the serial was successfully recognized and succeeded, I used the updater to try it out, and the aforementioned blocks went from four to two. It seems to have been rewritten, and XOR is no good.

I also don't know what $0070-$007F is doing.

Here's what I've got for now. fzabkar, thanks. Appreciate your cooperation.

« Last Edit: June 15, 2024, 03:47:33 am by squadchannel »
 
The following users thanked this post: feedback.loop

Offline asis

  • Frequent Contributor
  • **
  • Posts: 285
  • Country: ru
Hi,

Great job.
Congratulations.

-

In the update (in the body of the flasher) there is evidence that the scope can be updated to work with BUSHEALTH options (listening to various bus).
At least this trick worked on FLK199C to -> FLK225 v.804 and may be useful.
Also, expanded language support has been activated.
The way to activate it is to edit FlashTool.ini before the SW update program flasher is launched.
Now that you feel confident, you can afford it.

Good luck!
 

Online squadchannelTopic starter

  • Frequent Contributor
  • **
  • Posts: 423
  • Country: jp
  • deepl translate user
Hi,

Great job.
Congratulations.

-

In the update (in the body of the flasher) there is evidence that the scope can be updated to work with BUSHEALTH options (listening to various bus).
At least this trick worked on FLK199C to -> FLK225 v.804 and may be useful.
Also, expanded language support has been activated.
The way to activate it is to edit FlashTool.ini before the SW update program flasher is launched.
Now that you feel confident, you can afford it.

Good luck!

BUSHEALTH is also present in THS3024/3014. It is blocking the function somewhere.
Perhaps it is blocking between $0070-007F, or a calibration block. I'd like to try if I have time.

AS-i, RS-232, RS-485, Foundation(?), CAN, Ethernet, Modbus, Profibus




 

Offline patpat

  • Contributor
  • Posts: 30
  • Country: us
Re: [Solved!]Looking for firmware dump for Tek THS3024.(How to repair firmware)
« Reply #35 on: September 06, 2024, 01:08:54 am »
Awesome job.

To understand the memory map of these scopes I found useful the "Instrument Security Procedures" documentation from
https://www.fluke.com/en-us/support/instrument-security
I'm uploading here the ones for models:

Fluke 192B, 196B, 199B, 192C, 196C, 199C, 215C, 225C
&
Fluke 190-062, 190-102, 190-104, 190-202,190-204, 190-502, 190-504 (Series II)


These documents also say things like:

Quote
Fluke 192B, 196B, 199B, 192C, 196C, 199C, 215C, 225C
Security Summary:
The operating code (instrument firmware) stored in D101-D102 on the A201
module can be read using special remote interface commands. The instrument
firmware can be loaded using a dedicated Fluke software program.
Calibration constants stored in D101-D102 on the A201 module can be read
using special remote interface commands
. The calibration constants are
generated when the meter is sent through its calibration process and are
fundamental to the test tool operation.

Fluke 190-062, 190-102, 190-104, 190-202,190-204, 190-502, 190-504 (Series II
Memory Cleaning Instructions:
The operating code (instrument firmware) stored in D5000, D5002 can be read with dedicated remote interface commands (only available for use by Product Development).
The instrument firmware can be loaded selectively with firmware downloads that are distributed through the Fluke website. Fluke Service Centers can load firmware and configure the instrument with Service tools.
Caution: The upgrade tool will erase all saved screens and instruments setups.
Calibration constants stored in D5000, D5002 can be read with special remote interface commands (only available for use by Fluke Product Development or Fluke Service Centers). The calibration constants are generated when the test tool is sent through its calibration process and are fundamental to the test tool operation.

Off course these commands are still undocumented but looking at the Fluke/Tektronix 190-204 Firmware Upgrade it could be one of:

Quote
AS - AUTO SETUP - AS<cr>
AT - ARM TRIGGER - AT<cr>
BA  ---> BATTERY??
CI
CL
CM - CLEAR MEMORY - CM<cr>
CV - CPL VERSION QUERY - CV<cr>
DS - DEFAULT SETUP - DS<cr>
EM
EO
EX
FM
GD - GET DOWN - GD<cr>
GL - GO TO LOCAL - GL<cr>
GR - GO TO REMOTE - GL<cr>
ID - IDENTIFICATION - ID<cr>
IS - INSTRUMENT STATUS - IS<cr>
PC - PROGRAM COMMUNICATIONS - PC <baudrate> <cr>
PS - PROGRAM SETUP - PS [<saved_setup_no>]<cr>
QC   ---> QUERY CALIBRATION??
QF   ---> QUERY FLASH??
QI
QM - QUERY MEASUREMENT - QM<cr>
QP - QUERY PRINT - QP[ <screen_number>,<output_format>[,<block_transfer>]]<cr>
QS - QUERY SETUP - QS [<setup_no>]<cr>
QW- QUERY WAVEFORM - QW <trace_no>[,V|S] <cr>
RB  ---> READ BATTERY??
RC  ---> READ CALIBRATION??
RD - READ DATE - RD<cr>
RI - RESET INSTRUMENT - RI<cr>
RS - RECALL SETUP - RS <setup_reg><cr>
RT - READ TIME - RT<cr>
RW
SO - SWITCH ON - SO<cr>
SS - SAVE SETUP - SS <setup_reg><cr>
ST - STATUS QUERY - ST<cr>
TA - TRIGGER ACQUISITION - TA<cr>
VE
WB
WC  ---> WRITE CALIBRATION??
WD - WRITE DATE - WD <date><cr>
WT - WRITE TIME - WT <time><cr>
WW
RN
HO - HOLD - HO<cr>
CD
DT
RP - REPLAY - RP<cr>
IM
CN
PW
CC  ---> COPY CALIBRATION??
KY
PV
QV
IP
IV


Best,
Pat
« Last Edit: September 06, 2024, 08:30:14 pm by patpat »
 
The following users thanked this post: coromonadalix, feedback.loop

Offline feedback.loop

  • Frequent Contributor
  • **
  • Posts: 262
  • Country: us
Re: Looking for firmware dump for Tek THS3024.
« Reply #36 on: November 27, 2024, 11:56:15 pm »
...
We call them Upper/Lower with the BNC input on top. Both can be dumped using the TL866 and an adapter.
...
For correct recognition, there is a checksum at the end of each block. The checksum values must be matched correctly.
There are two types of checksums,
The result of adding the bytes from $0000 to $1FF7 (in the binary range of the block) goes into $1FF8,
The result of XORing the bytes from $0000 to $1FFB (in the binary range of the block) (including the result of the addition) is placed in $1FFC.


Thank you very much, squadchannel! This helped me to fix a Fluke 190-202 scopemeter.
You call the chips upper and lower, which may be a bit misleading. I would call them according to low (least significant) and high (most significant) 16 bits of the 32-bit data bus. The upper in your terminology holds the low 2 bytes.
How do you know about the checksums and XORs? I would love to understand how that should be done correctly.
Even better would be to get Fluke flash utility, so we wouldn't have to desolder the chips.
 

Offline fzabkar

  • Super Contributor
  • ***
  • Posts: 2804
  • Country: au
Re: Looking for firmware dump for Tek THS3024.
« Reply #37 on: November 30, 2024, 03:02:24 am »
How do you know about the checksums and XORs? I would love to understand how that should be done correctly.

Actually, I was the one who uncovered the checksum algorithms. Firstly, I noticed that there were several blocks of code which were damaged in such a way that they could be reconstructed by merging two halves. I had already written a tool for this purpose.

Next I noticed that each block had a lonely byte at the end, and this led me to suspect that these were checksum bytes. I then ran my CHECKSUM.EXE tool against these blocks. This tool tests a block of data using several checksum algorithms. It found a simple 8-bit sum plus an 8-bit XOR. You should find these tools attached to one or more of my posts.

If you have any blocks of code or data with an unknown checksum, I'll see if I can help you locate it. It would probably be best to start your own thread, though.
 

Online squadchannelTopic starter

  • Frequent Contributor
  • **
  • Posts: 423
  • Country: jp
  • deepl translate user
Looking for firmware dump for Tek THS3024.
« Reply #38 on: November 30, 2024, 03:18:51 am »
Yes, thanks to fzabkar. :-+

currently coding a utility tool for Scopemeter. It will take some time yet. :phew:
« Last Edit: November 30, 2024, 03:24:18 am by squadchannel »
 

Offline asis

  • Frequent Contributor
  • **
  • Posts: 285
  • Country: ru
Re: [Solved!]Looking for firmware dump for Tek THS3024.(How to repair firmware)
« Reply #39 on: November 30, 2024, 09:48:06 pm »
Hi,

[/quote]

You call the chips upper and lower, which may be a bit misleading. I would call them according to low (least significant) and high (most significant) 16 bits of the 32-bit data bus. The upper in your terminology holds the low 2 bytes.

[/quote]

@feedback.loop
The error was made in document 190BC___mveng0000.
I offer my representation of the A201 module circuit.
-
The circuit is very similar to the THS3024.
 


Share me