I was troubleshooting a counter I recently acquired and as part of that I had removed all the socketed chips to clean and re seat the chips.
That gave me the opportunity to save the firmware.
BTW my issue was resolved by re seating the ASIC chip.
Correction - that did not help the ASIC error after all. The unit needed a recapping as all the blue Phillips branded caps where significantly out of tolerance with high impedance.
In circuit they measured ok but out of circuit they where bad. The ceramic caps probably where making the in circuit readings acceptable.
The 33uf measured about 1...4 uf and 50 to 100Ohm
330uf measured 50...100uf and 4..10 ohms and a few 10uf measured 3uf and 10ohm.
After replacing all the caps no more issues with ASIC error on about 100 boot up sequences tested . Before the error would show up if the unit was in standby for more than 5s.
SRAM battery was completely dead but that did not cause any issues other than resetting settings after power cycle. Original BR2032(200mAh) replaced with BR3032 (500mAh)as I had it from another project.