Hi
I say again without further detailed information it is impossible to give precise help to you, a photograph of a scope without any setting information is of no use at all. This looks like an AM, vastly over modulated, RF signal with a 10Hz oscillation or a 1GHz oscillation modulation frequency that is being shown, we have no way of telling. DO YOU HAVE AM MODULATION TURNED OFF ?
With the Marconi 2022 series the first thing to check is the power supplies for correct voltage and ripple/noise, the second is to reseat all of the board interconnects, ribbon cables with DIL headers, they go intermittent or open circuit with time (oxidise), this is very important on the E variant for some unknown reason.
I have attached a scan of the 2022A AB1 RF processing board schematic, It MUST be used to fault find a 2022A, the main differences between the original 2022 up to the 2022D are mainly on the AB1 board, the changes were made as the maximum RF output power was increased upwards between variants. The 2022E was actually a cost cut down 2022C variant for the military, it lacks a high precision 10 MHz standard.
I have also attached a scan the relevant section of the service manual text that describes the operation of the AB1 board.
My best guess without any further detailed information is that you have either a control logic fault or a ALC loop problem, you must follow the signal paths from the IC8, IC9, IC10 signal path splitters through to the RF output amplifier to determine where the problem originates. For you information OM345’s usually fail totally, very rarely do they oscillate or go noisy, but it is not entirely unheard off.
At this time I can do no more without better information from you.
George G6HIG Dover UK