I have a spectrum analyzer Marconi TF2370 not working, power on is ok, but it certainly has some problems, there are known faults in the power supply?
Title: Re: Marconi Spectrum Analyser TF2370
Post by: Andy Watson on April 18, 2017, 03:57:59 pm
Are you aware of this group:

https://groups.yahoo.com/neo/groups/marconi_test_instruments/info
Title: Re: Marconi Spectrum Analyser TF2370
Post by: gergard on April 18, 2017, 05:29:30 pm
No, thanks for the information
Title: Re: Marconi Spectrum Analyser TF2370
Post by: gergard on April 19, 2017, 01:45:00 pm
The defect is, graticule ok, line present, noise none, scan locked, output square vawe 10MHz (25) is present, amplitude1,5Vrms, counter is on but locked at 04800000.0
If I connect 10MHz out (25) to the input (24), set sweep in manual mode and move bright line position, I see something.