Author Topic: Tektronix 576 HV Power Supply troubleshooting HV Probe Recommendations  (Read 2094 times)
Offline Johnny10Topic starter
I recently acquired a Tektronix 576 Curve Tracer B28xxxx serial number I believe from late 60's.
The CRT image flashes and dims and sometimes drifts after initial warm-up.
I checked all LV voltages and all measure within spec. and stay in proper range throughout my problem.
The 225 VDC measures from 199VDC through to 177VDC.
I have Fluke 87 for measuring voltages and cannot measure 4000 volt circuits.
I need a recommendation for High Voltage probes to measure CRT circuits.
One for DMM and one for Analog Oscilloscope.
Maybe something in $50 dollar range if possible.
« Last Edit: January 10, 2017, 03:06:40 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
Offline PaulAm
Re: Tektronix 576 HV Power Supply troubleshooting HV Probe Recommendations
« Reply #1 on: January 10, 2017, 07:00:16 pm »
If 6KV is good enough, there's a Fluke 80K-6 on ebay for $30 bin right now.  There's also one at $47 with a make offer.
If you need more than 6Kv, look for a Fluke 80K-40 which is good for 40Kv.  Either will work with most DVMs
At least if you're in the US.
The following users thanked this post: Johnny10
Online David Hess
  • DavidH
« Reply #2 on: January 10, 2017, 08:08:17 pm »
I have no complaints about the Fluke 80K-6 which I bought off of Ebay for exactly this sort of application.  The CRT cathode voltage is relatively low impedance and generally between -2000 and -4000 volts so the 75 megohm input resistance of the inexpensive 80K-6 is not a problem.  If you want to read the voltages from other CRT elements, then the 1 gigohm input resistance of the 80K-40 or 80K-15 will be better and of course the higher voltage probe is needed to measure any PDA (post deflection acceleration) voltage but this is seldom necessary.  These probes expect a voltmeter with a 10 megohm input resistance for accurate measurements.
A high voltage oscilloscope measurement is very rarely needed for CRT work but if you must, then the best options come down to a used Tektronix P6013 or P6015 which include shunt terminations.  These probes can also be used with a voltmeter if a 1 megohm termination is present.
Beware that new x100 high voltage probes that have a 100 megohm input resistance do not include a shunt termination and cannot be used safely with AC coupling without risking damage to the oscilloscope.