Author Topic: Meratronik V543 Voltmeter  (Read 834 times)
Offline mldevwTopic starter
« on: February 26, 2021, 09:03:02 pm »
on the plus side I just aquired an old nixie voltmeter from Meratronik with good documentation.
Now I figured it has a offset of about 160 mV in the 10 V measurment range,but I can't calibrate it, as the total potentiometer travel only does a 10 mV change.
I also tried calibrating the 1V range there is only a 5 mV difference but changing the potentiometer does absolutely nothing there.
(Yes, I made sure to warm it up and turn the right potentiometers, according to documentation).
Anyone had this fault before?
What point can I start analysing the issue?
Thanks in advance.
Cheers,
mldevw
Offline doktor pyta
Re: Meratronik V543 Voltmeter
« Reply #1 on: February 27, 2021, 12:13:07 am »
Please tell us readings for shorted input. Go through all DCV ranges.
We must distinguish zero error (offset) from gain error (slope).
10-turn potentiometer on the front panel 'zero' can have dirty wiper contact.
You can test it using ohmmeter.
You can do simple test - remove DC amplifier board and short it's output to ground (LO).
You should get stable  zero reading unless there is a fault in AD converter board.
P.S. I have a calibrated V543 but in other location.
If You won't solve the problem I can measure some voltages in the circuit for You - to have the reference.
P.S.2. Manual is there: https://www.google.com/url?sa=t&rct=j&q=&esrc=s&source=web&cd=&ved=2ahUKEwjJmYiT_4jvAhVlkIsKHQjKCS0QFjAAegQIAhAD&url=http%3A%2F%2Fbee.mif.pg.gda.pl%2Fciasteczkowypotwor%2FPolskie%2520mierniki%2FV543_2.pdf&usg=AOvVaw0wWdLV4loh_KVyHdZhIFXc
« Last Edit: February 27, 2021, 02:16:09 am by doktor pyta »
Offline bob91343
« Reply #2 on: February 27, 2021, 02:07:23 am »
A circuit diagram certainly would help.
Online coromonadalix
« Reply #3 on: February 27, 2021, 03:19:19 am »
what  i've found,  a djvu file  converted to pdf
from :  http://multimetry.tzok.eu/multimetr.php?typ=V543
Offline bob91343
« Reply #4 on: February 27, 2021, 06:16:47 am »
Always check the power supply first.  Voltages and ripple.
« Reply #5 on: March 01, 2021, 05:39:48 am »
Hello,
unfortuanetly I currently can't work on it, but I figured it is a slope error.
I will try to take out the DC amp board to check if there is a fault on the AD converter;
My zero potentiometer is fine.
Thanks a lot.
Cheers,
mldevw