Hi,
I've just acquired a Metrix MTX3250 multimeter, and I'm very happy with it. There is only one strange issue that I can not understand or resolve. Hopefully you can help me with this.
When I put the multimeter in Ohms mode and the test leads are disconnected, the multimeter shows 000.00 Ω. I would have expected to show .OL MΩ. When I connect the test leads from the MTX3250 to another multimeter and put that other multimeter in VDC mode, the MTX still reads 000.00 Ω. The other multimeters shows 5.6077 VDC. When I change the MTX from auto range to manual range, I can find the expected value of 10.926 MΩ. The other multimeter then shows a value of 0.6142 VDC. When I step down one range, the MTX3250 shows .OL MΩ and the other multimeter shows 1.1034 VDC. All other ranges show 000.00 kΩ, 00.000 kΩ, 0.0000 kΩ and 000.00 Ω. For all these ranges, the other multimeter reads between 5.1113VDC and 5.608VDC. When I have the correct reading in manual mode and switch back to auto ranging, the reading remains correct. Continuity seems to be working fine on the MTX3250.
Is this maybe a software issue? The current version of the software is 1.05.