| Keno1981:
hi i need some help, i have mettler pj400 and mettler pj 3600 which both were ok thrn after transporting them packed very well both show the same problem. they turn on but the 0.00g keep blinking. i also press the button to access the menu and again nothing happens. i opened and checked all sockets but everything is in place. what is your suggestion? i need help please! thanks |
| wraper:
Most likely "well packed" was not well at all with cone where platform is put on left with no protective cover or platform not removed at all. Look for broken thin vertical metal links between moving parts in area shown on the photo. I suspect the part where platform is put on is totally loose in vertical direction. |
| wraper:
Also creating multiple threads about the same thing is not welcome and will get attention from moderators. |
| DavidAlfa:
Don't make multiple posts for the same thing: https://www.eevblog.com/forum/repair/need-help-for-repair-or-fix-mettler-balance-scale-pj-400-and-3600/ |
| Keno1981:
sorry but i thought i write in wrong place first time. i will check and delete as soon as i find it.thanks |