My Fluke 5440B DC Calibrator has been slowly self-destructing. Here are the (many) symptoms:
1) Let it run for 4 hours, then do internal calibration. Immediately after the calibration returns, I set it to output 0.0 volts and the output measures about 12 uV.
I would have expected it to be much closer to 0 volts. I mean it supposedly just cal-ed itself!
2) After the four hour run and internal calibration, the unit never really stabilizes and will drift (downwards) about 20 uV over the next couple of hours.
3) At zero volt setting, the unit will occasionally trip to standby with an "Output Limit Fault, Output Under Voltage" error. Resetting the unit makes it fine again.
The voltage can swing to -1 or -2 volts for a second or two when this happens.
4) After many hours, I can sometimes get random errors, such as "Board Ack Fault, Check Pre-amp Board." Again, resetting it makes it fine again.
5) Internal analog tests show all pass even right after any misbehavior.
6) After a few hours at 1.0 volts, the unit will suddenly drop to about 0.953 volts (and at 10.0 volt setting, 9.53 volts) and stay there. I am graphing the output
on a Keithley DM6500 and I can see the drop is instantaneous (to the sample rate of the DM6500). Resetting the unit brings it immediately back to 1.0 volts,
until it happens again in a while.
I first thought maybe the ovens weren't working properly, but I monitored the various test points on the three ovens, and they seem to be acting properly.
My next instrumentation will be to simultaneously monitor the 1) reference, 2) DAC and 3) output when I am in condition #6 listed above to see what is
actually changing.
But does anyone have a guess for #6? What could cause the thing to drop from 1.0 to 0.953 suddenly and then recover immediately upon reset?