The Fluke 45 perform two consistency checks on the data of the eerom. One is the parity check, every cal constant must to be parity or odd depends of the rule. And the other check is a Checksum, in fact two checksum, one for cal constants (error 7), and another for system data (error 6).
If the checksum is corrupt. Just need to do a new write by the fluke. For example, go into calibration mode and do the calibration just of resistors. It´s easy with this DMM becouse you can select the value of the resistor to check. You must to check the manual, but for reference you take resistors for the 6 ranges (10ohm, 100ohm, 1kohm, 10K, 100K, 1Mohm) just read with your fluke the value of this resistors (or better with another DMM more accuracy) and you can use this resistors to do a self calibration. The gol is that you can select on the fluke that your reference resistor is a 1004.5 ohms intead a 1000.000 ohms.
This will fix a checksum corrupt (error 7).
I didn´t recalibrate my fluke. Just isolate and fix the corrupted data. Bought some cal resistors. And did a resistor calibration for fix the checksum.
I bought later a voltage reference to check, and it "appear" to be in tolerance. More than really I need.
I played a lot writing and re-writing the fluke eerom. And learned a lot of this system. But 18 months later my volatile memory is with other things....
Please, send me your list of cal constants. I can cross info and check is your offset can be wrong or not.