Hi...I have had an EDC MV106 operating for about 3 years and recently it failed. The odd thing is that it appears to be in the first decade switch assembly somewhere. My logic is as follows. Power supplies are good. Voltage reference appears good. The voltage out is approx 0 when the first switch is set to 0, About 2V when set to 1V, Overload (~13V) at 2V, about 3V when set to 2V. Every even number gives an overload. Every odd number gives about +1 volt than it should. ??!??? One other odd thing. The power switch is also open circuit and I had to bypass it to get anything to work. I don't believe in coincidences in troubleshooting...so... Anyway, I can't find a front panel schematic for the MV106 and was hoping someone out there would have one or at least can clue me in on how they use only 6 resistors for each decade. I like to avoid removing the rotary switch if possible.
All other decades appear to work as expected. I'm pretty sure the amplifier is good because of this.
My gut tells me there is enough information to find culprit or at least narrow things down.