Author Topic: Need help adjusting the Gw APS-9501  (Read 205 times)
Offline Alex06
  • Country: ru
« on: December 09, 2019, 10:32:08 am »
Hello.
I have AC POWER SOURCE APS-9501. For a long time he served me. And at one time, he took the testimony of U and I on the ADC, although the output in tolerance was measured by an external device.
I took the instructions and found in it the adjustment (calibration) of the ADC readings.
I managed to make an indication of U, but I didn’t.
I don’t know how to configure which buttons.
Need help.
Offline Alex06
  • Country: ru
Re: Need help adjusting the Gw APS-9501
« Reply #1 on: December 09, 2019, 10:48:31 am »
Offline Alex06
  • Country: ru
« Reply #2 on: December 09, 2019, 11:06:24 am »
Offline Ordinaryman1971
« Reply #3 on: December 11, 2019, 06:21:09 pm »
So, if I'm correct you are trying to calibrate the readings on it. How much is it off? Is the error consistent over the all the frequencies?
I have the same one, let me check the readings on mine.
So, I understand you are going through the calibration and you stumble upon some kind of trouble?
The following users thanked this post: Alex06
Offline Alex06
  • Country: ru
« Reply #4 on: December 12, 2019, 06:20:53 pm »
Hello.
Everything is written in the instructions. Section calibration.
My reading of l does not match.
For example.
~ 220V 50Hz., By overload.
I have a current of 1A.
The standard (external) shows 1A, and on the APS device itself 1.2A.
Offline Alex06
  • Country: ru
« Reply #5 on: December 12, 2019, 06:33:11 pm »
Can you change the current reading I up or down.
If possible, write
« Last Edit: December 12, 2019, 06:44:37 pm by Alex06 »