Hi!
I recently bought a broken Tektronix 222A mini scope. You turn it on and it beeps. But the CRT stays dark. At a cold start it draws about 583 mA, within 30 seconds it settles down to 552 mA at 12V. No battery is installed. It's powerd via the external power connector. I use the Tek 222 service manual, couldn't find anything useful for the 222 A model.
Link to the manual:
http://www.ko4bb.com/manuals/index.php?dir=Tektronix/Tektronix_-_222The battery charger SMPS seems to work fine (sits on the main board). It delivers the 9,8 V, as mentioned in the manual. The CPU board seems to work, too. I can measure the X/Y signals, also the Z signal is there.
The fault at my unit, I think, is very obvious at the A4-Power Supply Board.
The pages 26 and 70 in the manual are quite useful. Page 26 shows a block diagram of the power supply section and a little bit about the theory of operation. On page 70 you can find the interconnect diagram of all boards.
On the A4 PSU/CRT board is a TL594CD PWM controller. It gets the DISPLYE signal that is required for the IC to start working. Could not figure out if the feedback circuit is working correctly (But I assume this is a very important thing
) . The FET's Q401 and Q402 are switching, but they are really hot (~90°C).
There is nearly nothing on the output rails Vcc, Vee and AVcc.
Here are the measured values - with respect to GND - on the CRT connector (with CRT removed):
1 - Grid: -269V
2 - Filament: -233V
3 - Cathode/Filament: -246V
4 - Focus: -162V
Measured resistance at the CRT filament pins: 6,2 Ohms
Anyone here who had a similiar problem, or can give me a good hint?