Author Topic: Need Help for Backup Calibration Data Keithley 2015  (Read 990 times)

0 Members and 1 Guest are viewing this topic.

Offline letron77Topic starter

  • Contributor
  • Posts: 24
  • Country: fr
Need Help for Backup Calibration Data Keithley 2015
« on: July 14, 2019, 02:09:56 pm »
Hello everyone

I am trying to fix a 2015 Kiethley that is stuck.
Its display reads "No comm link"

Before any intervention I decide to save the firmware and files of setting
I have already read several post very interesting on this site

I bought a TLC866 II
It allowed me to back up ODD and EVEN firmware

By cons I do not know how to save the data "instrument setup and calibration constant" which is on the eeprom 24lc16b (U136)

on this page it is about the parametric data backup of a k-2001 by realizing a "DIY I2C dongle" to connect to the TLC866:

  https://www.eevblog.com/forum/testgear/restoration-glory-of-keithley-2001-dmm/msg384347/#msg384347

Can someone tell me how to realize it?
Is the procedure the same on a 2001 and 2015?
Should it desolder 24lc16B?

in advance thank you for your help

Alain
 

Offline Samogon

  • Frequent Contributor
  • **
  • Posts: 457
  • Country: us
Re: Need Help for Backup Calibration Data Keithley 2015
« Reply #1 on: July 16, 2019, 07:54:40 am »
You have to desolder it. In circuit it will play fool.
 

Online macboy

  • Super Contributor
  • ***
  • Posts: 2254
  • Country: ca
Re: Need Help for Backup Calibration Data Keithley 2015
« Reply #2 on: July 17, 2019, 06:36:40 pm »
You have to desolder it. In circuit it will play fool.
Maybe, but maybe not.
I have read the calibration data from the I2C EEPROMs of my K2001 meters by holding the CPU in reset, and reading with a TL866. Only 3 connections needed: data, clock, ground. The CPU must be held in reset to prevent errors caused by two I2C masters on the bus.
 

Online coromonadalix

  • Super Contributor
  • ***
  • Posts: 5897
  • Country: ca
Re: Need Help for Backup Calibration Data Keithley 2015
« Reply #3 on: July 17, 2019, 08:24:18 pm »
you have already your thread           why double posting ?

https://www.eevblog.com/forum/testgear/need-help-for-backup-calibration-data-keithley-2015/msg2549211/#msg2549211



reported to admins
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf