Author Topic: Need Help Repair Tektronix TDS520B  (Read 10033 times)
Online nctnico
Re: Need Help Repair Tektronix TDS520B
« Reply #25 on: November 21, 2016, 02:53:15 pm »
It seems the test only is failed if the dot or the 7 segment display is blinking.
Offline chihaxinh
  • Country: vn
« Reply #26 on: November 22, 2016, 01:46:55 am »
Hi All
Sory my mistake . LED DS1 display .E not .e  ( it's uppercase letter ) , so i thing have problem like broken trace or something like that.
It's TDS520B , not have much electrolytics capacitor. All cap i replace by new cap, esr have around 2.2R .
It's have 11 cap 33uF/10V and 10uF /35V .
So next time i desolder NVRAM and firmware IC to backup ! If anyone have Firmware or NVRAM data can share to me ?
Many tks you!
Best Regards
Nam Nguyen Hai
Offline singapol
  • Country: sg
« Reply #27 on: November 22, 2016, 06:53:15 am »
See if you can do diagnostic tests here:
http://old.timmestein.nl/index.php?page=tek-tds520-repair
You can copy nvram data using programmer and download to new nvram with fresh lithium battery.
Or if not put in new nvram and calibrate manually:
2430 2430A, 2432, 2440: calibration and faults due to NVRAM ...
https://forum.tek.com/viewtopic.php?t=137131
or extreme recovery of old nvram by cutting and removing epoxy to get at battery terminal : >:D
https://hohlerde.org/rauch/elektronik/geraete/tek2440/index.en.html
Offline chihaxinh
  • Country: vn
« Reply #28 on: November 22, 2016, 07:09:03 am »
Hi Singapol
NVRAM keep calibration value . in Tektronix 24xx calibration process very easy with signal gen and power source . But with TDS 5xx it's not easy , we need high speed signal gen and some special equipment.
Repair it's not easy , in next time i decided to replace CRT by LCD display, but we need equipment finish booting.  :phew:
Offline David Hess
  • DavidH
« Reply #29 on: November 22, 2016, 07:27:25 am »
On the 24xx DSOs, the external calibration process is easy.  On the 24xx analog oscilloscopes, it is more difficult but still possible with just the proper signal sources.  I think all of the TDS oscilloscopes also require special software which was not generally available.
Offline chihaxinh
  • Country: vn
« Reply #30 on: November 22, 2016, 07:43:40 am »
Yes. TDS5xx have software but running on GPIB card very old version ( ISA ) , and windows 3.1 .
My friend have Xeltek 501S programer , first step need backup NVRAM and Firmware , on internet some body post repair guide have only TDS520 or TDS520A not "B" version.
Hope NVRam not dead ...  :-BROKE
Offline chihaxinh
  • Country: vn
« Reply #31 on: November 22, 2016, 03:52:59 pm »
New update
After replace all electrolytic caps in Acq board ! Ds1 display on top segment only , likely boot hang !
And i connect osc to lcd display, lcd display many horizone line - Like bad video ram !
Next time i disasembly front panel and replace all electrolytic caps
Offline PrecisionAnalytic
« Reply #32 on: January 08, 2018, 07:09:22 pm »
I replaced all electrolitics on the digital and acquisition boards by tantalum capacitors and it's all been good for more than 3 years.
Just make sure you wash the hell out of the boards after removing the old leaky capacitors - brush and lots of IPA are your friends here.
Cheers
What brand, type and specification of tantalum capacitors did you use?
All the best