Author Topic: Need help to repair Fluke 867B  (Read 1779 times)
Online eric.niuTopic starter
« on: March 10, 2019, 08:35:25 pm »
guys, I need some help to repair my Fluke 867B, it suddenly stopped working on OHMS/DT/CT test, you can see from the picture, on OHMS it shows numbers with nothing plugin, also shows 2.5V on DT, 0.0567uf on CT.  Short test leads on OHMS get some unstable readings. Clean knob doesn't change anything. DCV and ACV seem ok but I didn't fully test all the ranges.
Any advice?
Offline Diesel11
Re: Need help to repair Fluke 867B
« Reply #1 on: May 12, 2022, 11:17:46 pm »
Same exact problem what was the fix? How?
Offline shakalnokturn
« Reply #2 on: May 12, 2022, 11:22:22 pm »
Usually a good starting point is to check PCB for battery leakage and clean the rotary switch contacts.
After that either you get hold of the schematic and start thinking or reflow the ASIC under the shield...
« Last Edit: May 12, 2022, 11:24:08 pm by shakalnokturn »
« Reply #3 on: May 13, 2022, 10:22:43 am »
I can help you too, how much are you willing to spend?  >:D
Offline Spectrum
« Reply #4 on: May 13, 2022, 12:49:34 pm »
also soldering around this switch