Quick update, decided to measure some resistors around DAC/ADC and I found two open resistors.
R361 should be 10k, after desoldering shows 60MR+(limit of my DMM). <- connected to signal switch/multiplexer(U30) responsible for ADC portion of the circuit.
R331 should be 30k, after desoldering shows 60MR+ <- connected to quad opamp(U25) from DAC portion of the circuit.
It looks like we may have plague with these green precision resistors, because I have seen guy on youtube(Gerry Sweeney) having same issues with this type of precision resistors being open in these agilent PSUs.
So now I have to replace them, unfortunatelly don't have any resistors with me(moved to another country some time ago).
Will order kit of 50 value of 0805 resistors, replace, see if it works, then order some proper 0.1% low ppm resistors.
EDIT: Found another open resistor - R339 100k
And another one, not open, should be 10k but I read 11.74k which seems to me quite a bit off for 0.1% resistor...
Most of these green 0.1% 10k resistors show 9.99k, so I probably should replace all of them that show +/- 50R off.
EDIT2: Another suspicious resistors that I haven't removed but show slightly off:
MARKING | VALUE | MEASURED VALUE
R357 10k 10.13k
R334 10k 10.20k
R332 10k 10.10k
R329 10k 10.08k
EDIT3: plague continues on main power board as well - R390 100K is open...