after I got the 2nd LO of my R&S FSA fixed now, me found a further issue.
If the span is >= 5MHz everything is just fine. (see upper part of 1st diagramm)
But if i set it right to 4,9999 MHz width, i get some peaks at the lower end of the spectrum between 100Hz to 4,9985 MHz.
The peaks show up at 87,9 kHz / 175,8 kHz and 263,7 kHz. (see lower part of 1st diagramm)
I was able to determine the signal from the YIG-oscillator to get "dirty" when you set a span lower 5 MHz.
The dirty signal
Sweep clean
Sweep dirty
I have been thinking about this for 2 weeks now and I am a bit to lazy to check the 5 modules involved in the control loop.
Could it be a mixing product of the mixer right to the YIG ?
Or may the multiplier have spectral purity issues ?
Maybe someone knows which module could NOT be the source of the "noise" ....
hack on,
induzer