Hi guys,
First of all sorry if I make language mistakes, I`m german

So rescently I bought myself an used Tektronix TDS520 Oscilloscope sold as defective. I bought that nice device and today the german Post delivered to me.
I switched it on and the "Diagnostic Log" popped up showing all the errors. The things that failed were the "Aquesition Processor Interface" and the "Acuesition" as it was showing the FAIL next to it.
Using good ol`google a day before it came delivered to me my first plan was that the capacitors were old and had leaked out.
I took the cover off the scope and realized that the capacitory were actually replaced by somebody else and were in perfect condition.
So I started to randomly check for Voltages on the A10 aquesition board and found out that there was Voltage missing on like 10 electrolytic capacitors that I marked in the photo with red circles.
I used the beeper on my multimeter to trace where these capacitors were connected to the ribbon cables (Yes all caps were in parallel) and found out that all of them were connected to the power supply with some wires of the ribbon cable.
Then I actually had no schematic or service manual ad all.
So I started to take things apart, disconnected all the Wires of the PCBs and took a look at the PSU board. I googled its part number online (It`s a "zytek 22904701") and found a schematic of it with i could then measure more stuff on the PSU board. I came to the conclusion that nothing with the power supply was wrong.
While actually looking for the PSU schematic I found a service manual for the TDS520 (more below).
With my hope that the PSU was intact I started to put things back together and tried to power the scpe up.
There obviously was the same problem.
So having the service manual I started checking different voltages of J26 and J27 of the A11 board as the voltages were listed in the service manual. All of these voltages were fine, maybe a little bit out but still in the listed tollerance range.
I´m prette sure it is something with those missing voltages.
The little 7 segment display is just showing one bar (also see pictures).
I have all the basic tools like Multimeters, Function gen, PSU, JBC Soldering and Hot air station so I think a mechanical repair should be no problem. I do not have a second oscilloscope. I have the service manual and the PSU schematics.
I would kindly apprechiate if there is some Tek specialist here and can help me with the repair since this to me is still a nice scope.
Thanks for reading this long story