Hi,
Got an Agilent 33220A from a friend, he said it was tossed out and probably needed repair. Sure enough, it boots with a self-test failed 606 error code:
606 - Self-test failed; cross-isolation interface.
This error indicates that the cross-isolation interface between the main processor (U101) and Synthesis IC (U501) has failed, or that the synthesis IC itself has failed
The display comes on and menu is functional, but it complains about some PLL being unlocked (from what I gather that's from the add-on card), and the signal output is dead.
So, I started up by checking voltages at semi-random spots, and quickly found that the test points marked +15V and -15V on the analog side had just 3-4 volts. A bit of schematic reading later and further probing I found that the isolation transformer is most likely all good (I measured ~15V across each of the rectifier diodes CR1302, CR1303, CR1309, CR1310), but something around the voltage regulators for the +/- 15 rails was not good. The 5V rail looked good.
I poked some further and finally found that the filtering cap C1308 on the output of regulator U1305 was basically a dead short across, unlike all the other caps I measured (they were all in kOhm range).
I'm a complete repair noob so I'd just like to verify that the next logical step would be to desolder capacitor C1308 and see if it improves things?
I've attached the relevant part of the schematics as well as a few pictures from the board.