Of course some of them needed about an hour to "form" (not sure about English word - running capacitor near rated voltage from current limited power supply).
Dissipation factor of 0.5 seems large - can You check it on other meter (with DC bias preferably) or on scope to confirm?
Hi Michael, good advice, I'll try that. C6 is in precondition with it's 16 rated volts. After that I do some measurements.
Here my Hameg 8142 with modulation capapilty comes in handy. I can apply AC with DC bias, current limited up to 1 A in 2 quadrants (source and sink limited to positive voltage).
Fan looks like it's mounted the wrong way. I would rip it out completely, since the newer 3456As have it removed too. Makes it quieter and less dusty inside.
Yes, it seemed a little odd to me as well. There is some kind of housing or duct where the fan sits in (not shown in the pictures).
From the backdating section and Change #10, which adresses the fan, I interpreted there are changes in some heatsinks.
For now I just had a rough look. But it's why I'd like to see this service note 12.
At least I want to hear the sound. But yes, I also prefere it without fan.
There was a lot dirt accumulated on the fan and on te filter. The filter was completly black. After a ultrasonic bath it is white again

.
Another recommendation: Unless you are absolutely sure what you are doing, don't try to calibrate that meter.
Hello Dieter,
that is a concern of mine. I have a 34465A I could check it against. It is not sufficient for perfomance check or calibration but it's the best I have.
But I can take it into account and be aware of higher uncertainty for absolute measurements.
The big problem is the lag of sources. Only a couple of power supplys up to 30 V and a function generator.
It's a project or two for the future.
I do not agree with the calibration : I have seen on a lot of 3456A that the drift over some years could be very important. Another tricky problem with the 3456A is the null with shorted inputs : there is no way to trim it and in most 3456A it is out of tolerance. Source are leaky fets in the input switches and it is a tediuous work to identify the faulty one.
Thank you for the hint. At least the zero check is easy to do with a piece of copper wire.
Regards
Martin