I have Fluke PM3390A scope, the problem is the typical illumination incandescent light bulb is burned out (I suspect). I finally found the service manual from my storage HDD, read that 450 page behemoth like an 2½ hours to find out that the type of the bulb is specified with one line in somewhere between pages 100 and 400 as 28 volts type... great... Now I need to open the scope to see which type of socket it is using.
I found also the opening proceduce. I realised that the unit should have a hard front cover which is supposed to be in place while pulling the case out. Oh, well I opened the screws tried to pull and jerk the sheetmetal case with different styles without that "face on the floor method" described on the manual. After 15 minutes without success I did put the srews back in and the scope back to its place. I do not need a stinky reticle.

PS.
If you need to went, be go for it. Let see your bad repair day.