Author Topic: prodigit 3320.Need repair or calibration. Need a scheme.  (Read 1623 times)
Offline Alex06Topic starter
  • Country: ru
« on: November 05, 2019, 08:16:59 pm »
Hello. This is still a newcomer. I decided to ask you for help, as I myself couldn’t do anything yet.
I have an electronic load Akip-1301A 60V / 30A
 (foreign analog of Prodigit 3320 60V / 30A) copy of 100%.
The malfunction lies in the inaccuracy of the readings that I submit to her voltmeter.
From 0 to 19 V is not resp., But from 20 to 60 V goes.
There are several tuning resistors, but so far they do not help.
On the Internet I was looking for a service manual or circuit, I did not find anything.
Maybe someone will help with the scheme or tell me where to go next.
« Last Edit: November 06, 2019, 07:57:05 am by Alex06 »
Online wraper
  • Country: lv
Re: prodigit 3320 Ремонт. Нужна схема.
« Reply #1 on: November 05, 2019, 08:28:27 pm »
Пост на всём кроме английского могут удалить, так что советую писать на нём. Тем более что большинство местного народа не будет заморачиваться переводить. И это не копия, а он и есть. Весь акип это тупо смена шильдика.
Offline dom0
Re: prodigit 3320.Need repair or calibration. Need a scheme.
« Reply #2 on: November 06, 2019, 09:11:13 pm »
The auto-ranging of the meter may be broken.
  • Country: ru
« Reply #3 on: November 07, 2019, 05:55:00 am »
One person suggested that it should be calibrated for accuracy. In the service manual it is. But on the Internet, I did not find it.
Online wraper
  • Country: lv
« Reply #4 on: November 07, 2019, 02:19:28 pm »
Do you mean up to 19V it does not work at all or you cannot adjust? If inaccuracy, how much. It could be a fault, not calibration problem.
  • Country: ru
« Reply #5 on: November 07, 2019, 04:19:16 pm »
measured the reading
Offline dom0
« Reply #6 on: November 07, 2019, 08:24:59 pm »
So it should be accurate to +- 0.05 % + 2 counts, yours is up to 0.2 V out (15 V / 15.2 V), +1.3 % error.
Range switching is indeed done at 20 V. So the lower range needs adjustment, preferably to an reliable, accurate multimeter. Handheld 4 1/2 digit would probably be ok. There is probably one trimmer or 20-turn pot for each range.
  • Country: ru
« Reply #7 on: November 08, 2019, 10:33:58 am »
I twisted some resistance, and put in place. They did not help. Only the general adjustment range was found
Offline danilnik
  • Country: ru
« Reply #8 on: February 11, 2022, 06:19:30 am »
Hello everyone.  Can I bring this topic up?  Maybe someone got a service manual or calibration method for this device?