Author Topic: Oscilloscope Hameg HM507 help in identifying the damaged element.  (Read 867 times)
Offline rubennssTopic starter
I have a Hameg HM507 oscilloscope in which one of the channels has been damaged.
The 602I241R1U integrated circuit located on the AT-C50 board is damaged (there are two same such integrated circuits on the board).
I am unable to decipher its full name and thus find a store where I could buy it and repair the oscilloscope.
Could you please provide me with a full name?
I am attaching photos of what integrated circuit I mean.
Greetings from Poland.
« Last Edit: October 14, 2021, 04:22:51 am by rubennss »
Offline Manul
  • Country: lt
Re: Oscilloscope Hameg HM507 help in identifying the damaged element.
« Reply #1 on: October 14, 2021, 10:20:30 am »
It must be MCP602 in 8-TSSOP package. Good luck.
The following users thanked this post: rubennss
« Reply #2 on: October 14, 2021, 04:56:05 pm »
Thanks a lot for your answer, you are right.
Just got a service manual from R & SCH from Germany.
It is as you wrote.
Regards.
Offline fzabkar
« Reply #4 on: October 14, 2021, 06:19:20 pm »
thanks, it will be useful in the future :-)