Hi Sel,
I got this old Oscilloscope which seems to have a fault.
From your description of the scope working initially and the present failure -
old aluminum electrolytic capacitors in the power supply are a quite suspect. Minimal response to the position controls and focus being way off point with certainty to a power supply problem.
Post up a few photos of the power supply when you can. I have attached the power supply board layout and schematic for reference.
Also, let us know what you have available for a meter to measure voltages with. The PM3200 manual shows -1500 Volts to the grid/cathode and +210 Volts to the deflection amplifiers, so the voltage extremes encountered are known. This oscilloscope does not use a Post Deflection Accelerator and therefore does not contain 10 kV to 15 kV.
Caution is still advised, none the less so as not to be
The power supply schematic shows that the scope uses a DC to DC converter to generate the various DC voltages. Most, if not all five outputs are likely to be low. Please do not attempt to measure the -1500 Volt output,
have been
from attempting this.
If you are comfortable measuring the +210, +90, +12 and -12 Volt outputs, please report the results.
RF+ Tech