One the channel 1 input is damaged. A colleague of mine connected one of the probes to a current transformer and injected a high voltage into it. The probe was melted and charged but was easily replaced.
Now there's a permanent 280mV offset, when the probe is connected to a low impedance source and DC coupled, which doesn't go away after running the auto-calibrate routine. It seems like the offset is caused by 280nA leakage current. When the input isn't connected to anything, the offset disappears. When I connect my multimeter to the input, it measures 280mV on the voltage setting and 280nA on the current setting.
Has anyone else had a similar problem?
The input is screened with a metal cage, which will be difficult to unsolder so I haven't done anything, except remove the case.