Author Topic: Owon TDS8204/PeakTech 1300 repair  (Read 321 times)
Offline ra-maTopic starter
« on: March 28, 2024, 01:05:48 pm »
Hello, I got a PeakTech 1300 aka Owon TDS8204 oscilloscope from my father in law. Issue description is "stuck in boot screen".
To check if it's potentially the power supply, I connected the main board directly to my lab power supply, skipping the internal power supply: The board takes 3 amps at 5 volts. It starts, shows the PeakTech logo but nothing more. So exactly same behavious as with the internal power supply.
In the input area of channel 2 I found a burnt 50 Ohms resistor (see picture) which I already swapped with a new one, but no change.
I was checking around that area for issues, but can't find anything:
- NEC relays are fine as far I can tell (measuring resistance between pins 1 and 8, unsoldering the one close to the burnt resistor, applying 4.5 volts, it's closing/opening).
- The battery (MS621FE) only has 2.5 Volts. I added another one in addition with 3.1 volts but no difference.
Any ideas where to look? Could it be more a soft- or hardware issue? Unfortunaltely I can't find helpful material online, doesn't seem to be a well selling oscilloscope...
I don't know the boot sequence of the Owon. My Rigol does some relais clicking at the end of the boot sequence, the Owon does a click (sound like several relais are switching) at the moment it gets power. Afterwards nothing happens anymore...
Looking forward for any tips and ideas.
« Last Edit: March 28, 2024, 01:11:50 pm by ra-ma »
Offline m k
Re: Owon TDS8204/PeakTech 1300 repair
« Reply #1 on: April 02, 2024, 02:32:42 pm »
Maybe battery voltage was too low, settings disappeared and the machine doesn't know how to boot.
Advance-Aneng-Appa-AVO-Beckman-Data Tech-Fluke-General Radio-H. W. Sullivan-Heathkit-HP-Kaise-Kyoritsu-Leeds & Northrup-Mastech-REO-Simpson-Sinclair-Tektronix-Tokyo Rikosha-Triplett-YFE