Author Topic: CRT oscilloscope issue  (Read 853 times)

0 Members and 1 Guest are viewing this topic.

Offline frankX0Topic starter

  • Newbie
  • Posts: 2
  • Country: nl
CRT oscilloscope issue
« on: April 29, 2021, 08:19:09 pm »
I have a Leader LBO-310A oscilloscope which was working fine until recently.
It now produces "extra" lines which go from the actual beam position on the screen to a point in the middle of the CRT. Both X- and Y-channels work properly,
The photos show the scope in XY-mode, where I can freely move the position of the actual measurement, but with the addition of the extra lines.
I have the service manual with the schematics, which does not look over complicated. But I also cannot deduct which part is causing this failure.
Help is appreciated!
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17106
  • Country: us
  • DavidH
Re: CRT oscilloscope issue
« Reply #1 on: April 29, 2021, 11:13:10 pm »
That looks a lot like ripple on the supply voltages because the aluminum electrolytic capacitors have worn out.
 
The following users thanked this post: frankX0

Offline bob91343

  • Super Contributor
  • ***
  • Posts: 2675
  • Country: us
Re: CRT oscilloscope issue
« Reply #2 on: April 29, 2021, 11:13:35 pm »
Bad CRT maybe.
 

Offline frankX0Topic starter

  • Newbie
  • Posts: 2
  • Country: nl
Re: CRT oscilloscope issue
« Reply #3 on: May 01, 2021, 05:40:00 pm »
I would not call this DC (see attachment). Nor would I call this ripple btw.
The 47uF/315V elco's are gone.
Thanks for pointing me in the right direction!
 