I have already repaired this oscilloscope model, and even like that, I do not understand anything about your explanations
- you only give a small portion of the diagram, where is the service manual?
Personally, I have it in my archives, but think of the other forum participants.
- The power supply for this oscilloscope is made up of two primary AC / DC power supplies of 25V each, placed in series to provide 50Vdc. and an inverter supplied with 50Vdc which supplies all the internal power supplies of the oscilloscope.
Are both 25V AC / DC power supplies working?
You have changed the Zener diode V3212, where is it on the diagram?
Where are the point X30 and R6318.? Ok
This is my first problem. The voltage can only be regulated between 8 and 9.5V. I can't get to 5V.
Indeed, with the components we see on this partial diagram, the lowest adjustable voltage should be 8.57V....R6321 should be 100K and not 10K